I'm attempting to repair a Tek 2431L scope (almost identical internally to 2430A). It is displaying symptoms that have led me to follow the dead start procedure in this service manual
https://data2.manualslib.com/pdf7/159/15831/1583003-tektronix/2430.pdf?87a103b2d0d33d02ac48fe9682d9795e I have no or very little voltage coming out of the voltage regulators (410). This is my first post here so please excuse any rookie formatting mistakes. I don't know the best way to put images in here, so I'll reference the service manual pages.
I have tracked down the primary error to be in the creation of the +10 and -5v reference voltages found on page 410 grid J2. I have a perfect 10.00v coming out of U180 but it drops to 2v after the RC filter. The inputs to U900 are both in that low 2v range. I was thinking that this was indicative of a bad op amp, but I replace the chip (LM308) with a TL070CP and the issue was better (less of a drop) but still incorrect and some component started heating up horribly (Q279, Q479, Q779, Q879) and the overcurrent shutdown kicked in, so the replacement was either not adequate or there is a bigger issue downstream.
I have an ESR meter on order, but other than that, I was wondering if anyone could point me in a direction to look? I'm lost.
Thank you so much in advance, and if there's any other info I can give, please let me know.