Scoped at TP18(Ch1), CR130 anode(Ch2) and CR131 anode(Ch3). There is a signal but it does not look exactly as it does in the manual. Look at the pictures.
The signal is the same no matter which switch/button I press.
Also, I measured the resistance of the relay coils. I did this by measuring at the cathode of the diodes going in to the relay coil.
Cr141-Cr142 = 63ohm
Cr143-Cr144 = 72.1ohm
Cr145-Cr146 = 72.5ohm
Cr147-Cr155 = 73ohm
Cr148-Cr149 = 72.9ohm
Cr150-Cr151 = 4.31ohm
Cr152-Cr153 = 72.3ohm
Cr140-Cr154 = 72.6ohm
Most look good except Cr141-Cr142 and especially Cr150-Cr151. I don't know if this is by design or if a coil or one of the IC are shorted or burnt on the inside.
While at it I also did some diode measurements.
Checking the forward bias:
Cr140-142, Cr144-149 ,Cr154-Cr155 and Cr131 had a forward voltage of anything between 0.619V-0.638V.
Cr143, Cr150-Cr153 and Cr130 had anything between 0.457V and 0.599V.
Then checking the reverse bias:
Cr140, Cr142-Cr147, Cr149, Cr151, Cr153, Cr131 had OL displayed.
Cr141, Cr148, Cr150, Cr152, Cr154-Cr155 and Cr130 measured 1.095V to 2.576V.
According to the manual all of these diodes are the same, so it looks like some of the diodes are toast.
Anyways, Thanks with the help so far bdunham7 I really appreciate the help!