Author Topic: [Solved]Tek 2465A -8V -5V too low  (Read 1935 times)
Offline GalenTopic starter
  • Country: cn
« on: October 30, 2019, 11:37:50 am »
I got a Tek 2465A, no trace. Sometimes, trace appear, but always 1 line, even I input signal.
As first attempt, I measured J119 voltage, found the +10V is right to the point, and most others are also within limit, except the -8V and -5V, I got -4.7V and -2.9V respectively. As per the SM, should fix the -8V problem first. I checked the -13.1V (before regulated to -8V), it's -8.7V only.
Anybody in this forum, has the experience to fix this problem, your advice is very appreciated. Thanks a lot in advance.
Galen
« Last Edit: June 02, 2020, 12:15:34 am by Galen »
Delighted when problem fixed
Offline tggzzz
« Last Edit: October 30, 2019, 11:58:37 am by tggzzz »
  • Country: cn
Re: Tek 2465A -8V -5V too low
« Reply #2 on: October 30, 2019, 01:27:59 pm »
dear tggzz, thank you very much.  Calin in that post only say 3 caps were replaced, but not say which cap. Hope Calin can provide advice.
Anyway, thanks again.
Galen
Offline tggzzz
« Reply #3 on: October 30, 2019, 01:36:17 pm »
Galen
There is a lot of information available on this forum (use the search facility) and elsewhere.
There's no point in us wasting our lives repeating it - poorly.
Online med6753
« Reply #4 on: October 31, 2019, 03:14:18 am »
If you are going to pull the Inverter/Regulator boards to perform the repair you might as well perform an entire re-cap of the supply at the same time. You'll find guidance here:
https://www.eevblog.com/forum/testgear/tektronix-2465b-oscilloscope-teardown/msg1658102/#msg1658102
  • Country: cn
« Reply #5 on: June 01, 2020, 07:36:07 am »
Fixed it long time ago. just close this topic.
A rectifier diode, of the -13V, is bad. Used a multimeter measure it, can not find it's bad. After replace it, the -13V is right, then -8V also right. Scope fixed.
Offline tggzzz
« Reply #6 on: June 01, 2020, 08:02:52 am »
Providing the solution is a courtesy to those that helped, and helpful to those that read the thread later.
That is aided if the OP adds "[Solved] " to the beginning of thread title.
  • Country: cn
Re: [Solved]Tek 2465A -8V -5V too low
« Reply #7 on: June 02, 2020, 12:17:40 am »
Thanks tggzzz, added the [Solved]. I'm happy if it can help anybody.
  • Country: cn
« Reply #8 on: June 02, 2020, 12:20:30 am »
Thanks med6753, yes, already recapped it in the repair.