Author Topic: Tek 2465A bootup  (Read 1720 times)
Offline zapvssTopic starter
Tek 2465A bootup
« on: April 26, 2019, 03:50:14 pm »
Hi All,
My Tektronix 2465A scope always boots up in Diagnostic mode screen after power up.  I have to to push the "A/B Trig" button to get it out of Diag. mode then it works normally.
It passes all of the tests at boot and everything works just fine after pressing the A/B Trig button.
Is there a setting that can prevent this behavior and have the scope display the trace after passing the tests at boot time?
Thanks
Denis
Offline alpher
Re: Tek 2465A bootup
« Reply #1 on: April 26, 2019, 06:41:21 pm »
Are you sure that there are no test fail mesages? May have to tweak readout intensity ?
« Reply #2 on: April 26, 2019, 07:04:43 pm »
Hi Alpher,
No error messages and readout intensity is up high enough to read anything that might show.
Offline Andy Watson
« Reply #3 on: April 26, 2019, 07:10:33 pm »
Have you tried forcing it to go through its diagnositc routines? It might show-up something that is being hidden by the boot process.
« Reply #4 on: April 26, 2019, 07:14:23 pm »
Yes I have and all 6 tests pass.
Offline pbarton
« Reply #5 on: April 26, 2019, 08:31:20 pm »
On power-up, the scope automatically performs a set of diagnostic tests.
When the tests have completed, is the TRIG'D indicator on the front panel flashing?
Offline edavid
« Reply #6 on: April 26, 2019, 09:18:26 pm »
I think you just have to remove the CAL jumper:
http://www.wb0smx.net/?p=2447
Offline ArcticGeek
« Reply #7 on: April 27, 2019, 01:23:32 am »
+1 On the cal jumper.
If you remove the cal jumper it will likely boot fine.
« Reply #8 on: April 27, 2019, 12:24:30 pm »
I checked the CAL Jumper and it is connected to the "No Cal" side.  I tried removing it and the scope wouldn't boot up.
Offline MarkL
« Reply #9 on: April 27, 2019, 01:14:52 pm »
You might want to try clearing CYCLE MODE using EXER 03.  It's described here:
  https://www.eevblog.com/forum/repair/tektronix-2465a-troubleshooting/msg2028619/#msg2028619
This is from from the service manual page 6-18, under "CYCLE ERROR CLEAR".  Cycle mode forces the scope into diagnostic mode on boot until it is cleared, and if I recall correctly, even if there are no more errors left to display.
« Reply #10 on: April 27, 2019, 01:35:49 pm »
Thanks MarkL,
That worked like a charm.  The scope now boots up properly.
Thank You
Denis