TEK 465 troubleshooting
« on: January 25, 2021, 05:08:13 pm »
Tektronix 465 (S/N B276519)

I am trying to get an old Tek 465 working. I have been able to get Ch-2 working but Ch-1 will only show the trace at the very top of the display when using Beam Finder. Using the service manual, O-scope troubleshooting guide and forums I think I have isolated the problem to the Ch-1 Preamp section.

I have performed the following:
  • checked DC voltage and ripple on all rails minus 2k rail all rails DC values were good. Ripple less than 4 mVAC RMS except 110V rail was 11 mV RMS (55 mVAC p-p) and had an unusual blip (see pic below). not sure if this is normal.
  • used the Common-mode technique on TP322 and TP324. this moved the trace to the vertical center of the display (per the troubleshooting guide this indicates that the problem is in a stage prior to the Vert. Switch. section, which also tracks with the fact that CH-2 works but CH-1 does not).
  • used the Common-mode technique on TP141 and TP147. This had no effect.
  • checked DC values at the following test points:
         
    • TP322: +0.6135VDC
    • TP324: -0.8034VDC
    • TP326: +1.6321VDC
    • TP141: +8.529VDC
    • TP147: +7.975VDC
  • re-seated the following transistors:
         
    • Q132
    • Q134
    • Q322
    • Q322

I am at a loss on how to narrow it down further. Any suggestions on what to check next?

SIDE NOTE: I can see the Trigger level as expected when pressing Beam Finder and Trigger View, but I noticed that the trigger level pot is inverted (turning it "-" raises the trigger level and vice versa) on both channels. Is this normal behavior? I am inexperienced with analog scopes.

« Last Edit: January 25, 2021, 08:37:56 pm by beaug »
 

Offline Paul Moir

  • Frequent Contributor
  • **
  • Posts: 927
  • Country: ca
Re: TEK 465 troubleshooting
« Reply #1 on: January 25, 2021, 05:46:47 pm »

I can see the Trigger level as expected when pressing Beam Finder and Trigger View, but I noticed that the trigger level pot is inverted (turning it "-" raises the trigger level and vice versa) on both channels. Is this normal behavior? I am inexperienced with analog scopes.

Not really helpful but I happened to have a plugged in 465 next to me.  This is normal with DC coupling.
« Last Edit: January 25, 2021, 05:48:47 pm by Paul Moir »
 
The following users thanked this post: beaug

Online med6753

  • Super Contributor
  • ***
  • Posts: 11329
  • Country: us
  • Tek nut
Re: TEK 465 troubleshooting
« Reply #2 on: January 27, 2021, 03:32:17 am »
Since you have one channel working you can use it for comparison troubleshooting. Set both channel 1 and 2 to the identical setting on volts/div. Apply identical signal, such as the calibrator, to both channels. Trace the signal on both channels through the vertical circuits until you come upon a difference. That should be the area of the defect. 
An old gray beard with an attitude.
 