Thanks for the answers gentlemen. To rule out the HV tripler section, I rigged up an external tripler with 3 caps and 3 diodes. The HV voltage in the 465M is 10kV. And the rigged up tripler generates it. As for the rest of the caps and diodes in that U550, I had the working 365 (which had a working U550) also develop the same problem after I swapped the time base/horizontal. Yes, it is one big single unit with a couple of daughter boards on it. 465M has 4 major parts as I see it. The vertical, time base, power supply with the HV components on the main board and the tube assembly. So, when I moved the “parts” around, it was always a complete unit in the case of vertical and time base.
Right now, with the vertical and the time base unplugged, the fuse still blows if I connect the HV lead to the tube. I have also tried 3 separate tubes. This is the reason I get the feeling that the problem is in the main board (in addition to the time base that caused it-maybe). Like Med says, I may have to fabricate a new U550 as I did earlier (my previous thread) but I am not sure if that will resolve the problem. I say this because, the 2nd unit had a working U550 and that unit developed the same problem after I swapped the board. Unlike the 465, the 465M has an “integrated” HV module and the last time around it took me a while to rig the thing.
Thanks again.