some exta inforamtion:
I made same test measurement. The conclusion:
The instrument real max bandwidth is 3.5 Ghz!
If i let the bandwith limiter in HW mode and i use the channel without PI filter, the bandwidth is at least 3.5 Ghz!
I can see the 2.4 Ghz WiFI signal on the FFT without any down convertersion. After 3.7 Ghz the signal power start to decrease. Approx 3dD
The conclusion. Despite of difference of DPO7054,DPO7104 and DPO7254,DPO7354 acquisition board, the bandwidth capable is same.
In the original case when my flash was good (DPO7054), the 500Mhz bandwidth limit is limited the instrument capabylities.
But in the DPO7104 case only the 1Ghz(DSP) limit the bandwidth to 1GHZ, but the 1GHZ(HW) limit option not limit anything. You can use the instruments capabyliti.
I already attached many pictures, but my previous post lost.
I will ty again.
My plan: open again the whole instrument and read out again the flash.
Other success. I found a core2 motherboard wich can work with the original WinXP image. Only one motheboard. Other isn't work. I don't kow why. Core2 processor and ssd combination could increase the performance of orginal P4 + HDD combination.
I'm sill looking for a not corrupt DPO7104 flash content.