Hello everyone,
I recently purchased a broken Tektronix TDS 360 that has issues digitizing its input. At low sampling rates (<50ksps), the output is mostly correct except for in the middle and end of the sampling interval. At the very end, the signal appears very distorted or doesn't look like the input signal at all. In the middle, there seems to be some kind of phase shift error which results in the wrong part of the signal being digitized.


The phase error is consistent in its location though doesn't always occur.

At higher sampling rates (>100ksps), the signal becomes increasingly corrupted until it disappears completely at 1gsps.
Interestingly, the degree of corruption depends on the frequency of the input signal. A low frequency signal paired with
a low sampling rate and a higher frequency signal and fast sampling rate will result in similar levels of corruption, but a low
frequency signal with a fast sampling rate will be very corrupted.

The oscilloscope passes the power on self test but running the diagnostics utility results in the following errors.
diagAcq_fisoCell
diagAcq_acqMemAddr
I have been looking through the maintenance manual but it seems like repair = replace. I have checked the power supply
and all of the voltages seem correct. The main CPU/Acquisition board only has 3 electrolytics on it and seems to be in
excellent shape. I have been tracing out the signal path using the service manual and have concluded that the issues begins
somewhere between the analog front end and the A/D converter. I have been able to monitor the waveform at the
output of the sample driver and it seems to be correct. The waveform at the input to the A/D is this strange digitally modulated
version of the input. While I can't see the corruption on my analog scope, because of the general blurryness and instability
of this signal, I suspect that the errors are introduced before the A/D. I have also tried disconnecting the A/D from the signal
path, which causes the corruption to disappear. At this point I believe that the errors originate in the FISO sample memory,
which is this strange CCD analog memory device that is made of unobtanium.

Image I stole from this thread:
https://www.eevblog.com/forum/repair/pls-help-me-repair-tektronix-tds-380/
FISO driver output bottom, FISO output top. FISO signal appears to be a series of discrete analog voltage levels.

At high frequencies the FISO output doesn't make much sense.
Do you think there is any hope for this scope? Could it be something other than the FISO memory or something that is negatively affecting its operation? I don't want to replace the main board because the scope really isn't worth it, but I do need a digitizing scope for some of my upcoming projects. Do any of you have a lead on this unobtanium part?
There is one other issue that I haven't gotten to diagnosing yet because it seems to be separate from the sampling errors.
Sometimes, signals exhibit a stair step effect that causes them to shift up and down periodically. This happens even when
the inputs are coupled to ground and the steps appear to be consistent in height. The period is very slow, approximately 1
second. The stepping can't be detected at the output of the FISO driver or the FISO itself, so I'm not sure where it is coming
from.