Hi,
I bought a completely dead Tek TDS 540 (no suffix).
When plugged, it tripped the RCD, due to a leaky CY cap.
After replacement, no more RCD tripping, but the scope won't start.
Following service manual troubleshooting hints, I'm testing the power supply (119-3371-03) with no load.
The attached pic shows what I see across C12: don't be fooled by the Fluke readouts: the voltage is downscaled 100-1.
The average voltage is correct (around 410 V) ma where are those tremendous spikes (200 Vpp) coming from?
I'm stuck: please any help will be highly appreciated.
Thank you
Marco