I have had a 535A scope for many years and love it. Unfortunately, suddenly had a problem, then multiple problems. The ceramic connections of wires and parts, I shall call pins.
1. Initial problem was vertical, both A and B suddenly went off screen. I checked the supply voltages at the labeled check points on top of the scope and the -150 and +225 voltages were way off. Replaced tubes, cleaned tube pins but did not help.
2. Spent some 30 hours checking passive parts, did not find any problems. After an arc from a 500 volt supply pin to a grounded pin (on the ceramic/arm that swings out), I realized the problem could be leakage or worse between pins. So I cleaned (sandpaper and no residue spray) between pins throughout the scope, and the vertical started working properly. All the voltages measured fine, within 1-1 3/4% at the test points on top of the scope.
3. Checked the scope and Alternate and Chop do not work. Just a single line. The vertical position works fine, respectively.
4. I checked voltages at the CA plugin socket, no CA plugged in, and these were the voltages.
Pin 9 -50 volts
Pin 10 +131
Pin 11 +119
Pin 12 120
With the CA plug in installed.
Pin 9 -49 volts
Pin 10 +72
Pin 11 + 88
Pin 12 + 88
The vertical measurements seem accurate when using the Amplitude Generator Cal out.
When I set Horizontal Display to B, Channels A and B still work, but the horizontal line is only about 1/2 the width across the screen as when Horizontal Display is set to A. Alternate and Chop still do not work.
I checked and plate of V245 has a pulse. However, the screen pin 6 does not. V154 has nothing. I did spray clean the switches, but the spray leaves no residue.
I can use the scope in channel A or B, but sometimes it is nice to use Alternate.
By the way, I cannot find 100 ohm resistors R 603, 605, 607, 609. So difficult to find the parts.
I am an analog audio design expert, but not very good with scopes. What I am wondering is if the problems could point to one particular area or part, or whether there are multiple problems. Is this just a coincidence of multiple problems?
Thanks for any help and time/efforts.