| Electronics > Repair |
| Tek TDS7404 tips about log welcome |
| << < (4/7) > >> |
| charlyd:
yesterday i worked on an other very strange issue. I installed a SSD a while ago and it worked ok with the SF810 board. --- Code: ---/ no jumpers on the drive / bios auto detect / direct connected to the cable / without drive bracket / without conversion adapter [visible in the bios as] SSD as primairy master CDROM as secondairy master --- End code --- all worked well and OSses could be installed ok then i swapped to the newly bought SU810 board with the same config, because i want to end at winXP embedded with a very small footprint --- Code: ---/ no jumpers on the drive / bios auto detect / direct connected to the cable / without drive bracket / without conversion adapter [visible in the bios as] SSD as primairy master CDROM as secondairy master --- End code --- drives were detected as above so no changes were made to the bios drive settings and i started installing W2K, while installing i created a new partition and formatted NTFS. After the reboot i saw the windows 2000 logo for about 10 seconds and then i got stuck in 0x0000007B (INACCESSIBLE BOOT DEVICE), knowing my SSD is 100% ok. after troubleshooting for a very long time i figured out that a HDD worked ok and that with an SSD auto detect is not really auto detect in those days bioses. ( both drives HDD and SSD didn t have any jumpers at this time) so when you go and look inside the bios you see a visual and this looks good but is not! This is NOT enough. you have to force your SDD to be the master on the primairy bus. so after placing the right jumper on the drive and the visual doesn t change at all, it all start working. below you can see what needs to be done --- Code: ---/ jumper set to master on the drive / bios no change / direct connected to the cable / without drive bracket / without conversion adapter [visible in the bios as] SSD as primairy master CDROM as secondairy master --- End code --- so keep in mind that jumpering your SSD drive can solve a lot of problems in generally with SSD drives and (older) motherboards. and then installing Windows 2000 on a SSD was an easy job with a succesfull. i finished installing wimdows xp embedded pos. and did some tekscope timing. complete bootup 90sec and shutdown 25sec complete off. 1.0Ghz cpu. i am still working on getting a 1.4GHz /256/100Mhz Front side bus. But still no succes. |
| Guy_Ellis:
Hi Charlyd, Sorry for the delay in coming back to you, work has been very busy lately. Unfortunately the TDS7704 unit with the faulty trigger chip, might have another fault. SPC fails big time with pages of error messages. I recall that on the TDS6604 we fixed last year there were only a handful of error messages before we replaced the trigger chip. Replacing the trigger chips is not too difficult, but due to the board size you might need a second pair of hands. 1. Removal Apply gel flux to all pins and desolder with a good quality thin wick Once every pin is desoldered (confirm by probing each pin with a dental pick) put the PCB up on it's edge and heat the PCB directly under the chip with a largish tip in the iron (2-3mm). Use extra solder to improve heat transfer. After 20-30 seconds you should be able to remove chip with a pair of tweezers but keep the heat applied - this is where you need a second set of hands. 2. Cleanup Using plenty of flux and thin solder wick, remove all excess solder from pads and the area where the slug is soldered - you must get this nice and flat, and clean old flux off with a good flux cleaner. 3. Solder Apply flux and solder paste on the slug pad on the PCB. I use a low temp paste that melts around 180 degrees. Drop the chip on and solder two corner pins to hold it in place. Flip the PCB over and heat the PCB directly under the chip with a largish tip in the iron (2-3mm) for around 30 seconds. Again some extra solder between the tip and the PCB will improve heat transfer. Flip the PCB over again and confirm that the slug has soldered - push the chip with a small screwdriver and see if it moves. If it does then repeat the heating process. Once the slug is soldered, solder the rest of the pins and clean any remaining flux off when finished. Good luck - Guy |
| charlyd:
Hi ok, after some time i managed to change the trigger chips and yes they were both dead... especially the first one took me some time before it was in place how i wanted it and you have to warm up the area arround the chip more then you want but ok. after booting up the scope with a Fan blowing close over the ACQ board to keep the chips cool and copying in a complete CALcons directory content from an other scope my SPC still fails. From what i see in the log i have two problem areas still: ( i copied highlighted parts from the log block ) tips are very welcome, for sure what problem 2 can be?? -------------------------------Problem 1 ------------------------------------ extd chksum failed, stored = 0xffffffff calc = 0xdb766060 0x16975c0 (tShell): cc_init: extended bank17 chksum failed, NOT reset to factory 0x16975c0 (tShell): cc_chkLimits: CC_ACQTI_MAIN_ET_MAX was 162; reset to factory 0x16975c0 (tShell): cc_chkLimits: CC_ACQTI_MAIN_ET_MIN was 984; reset to factory -------------------------------Problem 2 ------------------------------------ minMain 2000; maxMain 0; minDelay 2000; maxDelay 0 total Acqs 100001; Acqs without change 0; fail metaMti 100000; metaDti 0; metaMAT 0; metaMUT 0; metaDUT 0 measureTIrangeAndOffset() failed minMain 2000; maxMain 0; minDelay 2000; maxDelay 0 total Acqs 100001; Acqs without change 0; fail metaMti 100000; metaDti 0; metaMAT 0; metaMUT 0; metaDUT 0 measureTIrangeAndOffset() failed ET cal failure (SLOW FAST) here is my complete log: --- Code: ---UUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUU romPost: MPC106 Walking-one passed. romPost: MPC106 configuration passed. romPost: First PCI access passed. romPost: PC87560 Walking-one passed. romPost: PC87560 configuration passed. romPost: DEC21554 configuration passed. romPost: PERFORMING ROM CHECKSUM TEST... romPost: ROM checksum passed. romPost: PERFORMING DRAM CELL TEST... romPost: Enabling data cache... romPost: Data cache enabled. romPost: Enabling instruction cache... romPost: Instruction cache enabled. romPost: DRAM Cell test passed. romPost: Disabling data Cache... romPost: Data cache disabled. romPost: Disabling instruction cache... romPost: Instruction cache disabled. romPost: PERFORMING DRAM MARCH TEST WITH CACHE OFF... romPost: Writing address to each *address... romPost: Reading address from each *address... romPost: DRAM march test passed... romPost: PERFORMING DRAM MARCH TEST WITH CACHE ON... romPost: Enabling data cache... romPost: Data cache enabled. romPost: Enabling instruction cache... romPost: Instruction cache enabled. romPost: Writing address to each *address... romPost: Reading address from each *address... romPost: DRAM march test passed... romPost: PERFORMING DRAM WALKING-ONE TEST... romPost: DRAM Walking-one test passed. romPost: PERFORMING NV-MEMORY WALKING-ONE TEST... romPost: NV-MEMORY Walking-one test passed. romPost: Disabling data Cache... romPost: Data cache disabled. romPost: Disabling instruction cache... romPost: Instruction cache disabled. romPost: Enabling instruction cache... romPost: Instruction cache enabled. romPost: POWER-UP DIAGNOSTIC COMPLETE. Hel VxWorks System Boot Copyright 1984-1996 Wind River Systems, Inc. CPU: Tektronix HPP - PowerPC 740 Version: 5.3.1 BSP version: 1.0.8.0 Creation date: Jun 27 2003, 09:52:32 Press any key to stop auto-boot... 0 auto-booting... boot device : sm processor number : 0 file name : c:/vxboot/vxWorks.st inet on backplane (b): 192.168.0.1:ffffff00 host inet (h) : 192.168.0.2 gateway inet (g) : 192.168.0.1 flags (f) : 0x1008 target name (tn) : tds7000 startup script (s) : c:/vxboot/topscript.hw other (o) : nvfs=0x1000 Initializing backplane net with anchor at 0x1e80000... done. Backplane anchor at 0x1e80000... Attaching network interface sm0... done done. Attaching network inPC1 dio device driver installed terface lo0... done. Attaching DIO client... done. Loading c:/vxboot/vxWorks.st...1773516 Starting at 0x200000... Host Name: bootHost User: target Initializing backplane net with anchor at 0x16e0000... done. Backplane anchor at 0x16e0000... Attaching network interface sm0... done done. PC1 dio device driver installed Attaching network interface lo0... done. Adding 6008 symbols for standalone. ]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]] ]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]] ]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]] ]]]]]]]]]]] ]]]] ]]]]]]]]]] ]] ]]]] (R) ] ]]]]]]]]] ]]]]]] ]]]]]]]] ]] ]]]] ]] ]]]]]]] ]]]]]]]] ]]]]]] ] ]] ]]]] ]]] ]]]]] ] ]]] ] ]]]] ]]] ]]]]]]]]] ]]]] ]] ]]]] ]] ]]]]] ]]]] ]]] ]] ] ]]] ]] ]]]]] ]]]]]] ]] ]]]]]]] ]]]] ]] ]]]] ]]]]] ] ]]]] ]]]]] ]]]]]]]] ]]]] ]] ]]]] ]]]]]]] ]]]] ]]]]]] ]]]]] ]]]]]] ] ]]]]] ]]]] ]] ]]]] ]]]]]]]] ]]]] ]]]]]]] ]]]]] ] ]]]]]] ] ]]] ]]]] ]] ]]]] ]]]] ]]]] ]]]] ]]]]]]]] ]]]]] ]]] ]]]]]]] ] ]]]]]]] ]]]] ]]]] ]]]] ]]]]] ]]]]]]]]]]]]]]]]]]]]]]]]]]]]]] ]]]]]]]]]]]]]]]]]]]]]]]]]]]]] Development System ]]]]]]]]]]]]]]]]]]]]]]]]]]]] ]]]]]]]]]]]]]]]]]]]]]]]]]]] VxWorks version 5.3.1 ]]]]]]]]]]]]]]]]]]]]]]]]]] KERNEL: WIND version 2.5 ]]]]]]]]]]]]]]]]]]]]]]]]] Copyright Wind River Systems, Inc., 1984-1997 CPU: Tektronix HPP - PowerPC 740. Processor #0. Memory Size: 0x2000000. BSP version 1.0.8.0 . WDB: Ready. usrNvFsInit: NAME=/nvram BASE=0xfd0ff100 BPB=512 BPT=8 NB=8 BO=0 nvfs: STATUS: Mounting existing NVRAM file system KERNEL: THERMAL CONTROL INITIATED Executing startup script c:/vxboot/topscript.hw ... dbFsIoctl: invalid code 3 ################### topScript starts ############################### cd "dbfs:c:/vxBoot" value = 0 = 0x0 ld < TekScope.o value = 23662048 = 0x1690de0 ld < TekScopeSM.o value = 23661408 = 0x1690b60 pDbDev = ramDevCreate(0, 512, 1152, 1152, 0) new symbol "pDbDev" added to symbol table. pDbDev = 0x1692720: value = 23668544 = 0x1692740 = pDbDev + 0x20 pDBfs = dosFsMkfs("/db", pDbDev) new symbol "pDBfs" added to symbol table. pDBfs = 0x16909a0: value = 23661056 = 0x1690a00 = pDBfs + 0x60 copy ("TekScope.DBF", "/db/TekScope.DBF") value = 0 = 0x0 overrideFPdefaults() value = 0 = 0x0 measurePPCtimebase value = 0 = 0x0 registerMemPartitionSizes ( 0x00000000 ) value = 0 = 0x0 semInitForQueueCheckLogMsg value = 21818112 = 0x14ceb00 initSysWaiter() value = 0 = 0x0 raxInit value = 0 = 0x0 paiInit done value = 16983892 = 0x1032754 = PaiBaseRegisterTable + 0x948 verifyAcqBoardID value = 0 = 0x0 tcsLoadTableToNVRAM "pc:c:/vxBoot/Setup.tcs" value = 0 = 0x0 updateAcqProcFirmware("dbfs:c:/vxBoot/acqProc.hex") 0x16975c0 (tShell): AVR FLASH contents match file dbfs:c:/vxBoot/acqProc.hex - n o programming needed. value = 1 = 0x1 cd "pc:" value = 0 = 0x0 raxInitializeDevices done 0x16975c0 (tShell): (Re)starting Acq Processor value = 21815208 = 0x14cdfa8 excHookAdd(stopStressForException) value = 16770764 = 0xffe6cc = stopStressForException dsyInit value = 21806440 = 0x14cbd68 dsyWPApplyNextRegs value = 16973768 = 0x102ffc8 = DsyRegisterTable + 0xab0 # remove '#' from next line to set powerup scope state to factory default: # NvramClearDb = 1 nvram_initFileSys value = -49409789 = 0xfd0e1103 = pDbDev + 0xfba4e9e3 TmgrInit("/db") value = 0 = 0x0 PhxProxyInit value = 20819072 = 0x13dac80 rtlInit value = 0 = 0x0 sysExtMemAddToPool memSysPartId,0x02000000,0x0e000000 value = 0 = 0x0 hwAccount_init sysEnetAddrGet: 93:ef:1f:87:c5:2a value = 0 = 0x0 GpibInit value = 267500128 = 0xff1ba60 wfmMemInit(1) value = 57781984 = 0x371aee0 vertInstall value = 57774948 = 0x3719364 calConStatus = cc_init(2) 0x16975c0 (tShell): Reading 300 bytes of service cal data from EEPROM[0] into 0x 3719220 0x16975c0 (tShell): nvMem_read succeeded 0x16975c0 (tShell): Reading 82 bytes of service cal data from EEPROM[300] into 0 x3719300 0x16975c0 (tShell): nvMem_read succeeded 0x16975c0 (tShell): extd chksum failed, stored = 0xffffffff calc = 0xdb766060 0x16975c0 (tShell): cc_init: extended bank17 chksum failed, NOT reset to factory 0x16975c0 (tShell): cc_chkLimits: CC_ACQTI_MAIN_ET_MAX was 162; reset to factory 0x16975c0 (tShell): cc_chkLimits: CC_ACQTI_MAIN_ET_MIN was 984; reset to factory 0x16975c0 (tShell): Optical hardware is not present. calConStatus = 0x1335528: value = 131072 = 0x20000 vertPostCalRestoreInstall value = 0 = 0x0 MathMeasCreateTheMathandMeasurementLibraries value = 57602368 = 0x36ef140 initAcqMapper 0x16975c0 (tShell): initAcqMapper: Enter value = 0 = 0x0 trigMapInit value = 0 = 0x0 dsyMapInit value = 0 = 0x0 mathMapperInit value = 0 = 0x0 utilMapInit value = 57599712 = 0x36ee6e0 TmgrHookup value = 0 = 0x0 PhxProxyHookup value = 0 = 0x0 hookupAcqMapper value = 20074648 = 0x1325098 = AcqMapperRtlDataList MathMeasMapHookup value = 23687912 = 0x16972e8 dsyMapHookup value = 23688016 = 0x1697350 utilMapHookup value = 23688048 = 0x1697370 rtlHookup value = 0 = 0x0 horizMapHookup value = 23688008 = 0x1697348 fastFrameMapHookup value = 23688032 = 0x1697360 trigMapHookup value = 23688040 = 0x1697368 vertMapHookup value = 23688032 = 0x1697360 spcMapHookup value = 23688032 = 0x1697360 autosetMapHookup value = 23688072 = 0x1697388 timeSyncMapHookup value = 23688064 = 0x1697380 auxTrigMapHookup value = 23688048 = 0x1697370 dpoMapHookup value = 23688040 = 0x1697368 TmgrStartup value = 56625168 = 0x3600810 PhxProxyStartup value = 56574624 = 0x35f42a0 taskDelay(2) 0x3681ea0 (tmgr): >> optNewOptKey: Bad instru Id value = 0 = 0x0 gpibHwInitC value = 0 = 0x0 gpibTaskSpawn value = 56270400 = 0x35a9e40 gpibIoInit "/dbios/DBIO005","/dbios/DBIO006" value = 56356992 = 0x35bf080 enablePowerupDiags = 0 enablePowerupDiags = 0x12fad08: value = 0 = 0x0 EnableHorizInterpolation=1 new symbol "EnableHorizInterpolation" added to symbol table. EnableHorizInterpolation = 0x14bc7e0: value = 1 = 0x1 runPowerupDiags done dipSwitch = 0x0, enablePowerupDiags = 0, puMode=0 CalDiagPowerUpDiagsEnabled = 0 !! SKIPPING POWERUP DIAGNOSTICS !! minMain 2000; maxMain 0; minDelay 2000; maxDelay 0 total Acqs 100001; Acqs without change 0; fail metaMti 100000; metaDti 0; metaMAT 0; metaMUT 0; metaDUT 0 measureTIrangeAndOffset() failed minMain 2000; maxMain 0; minDelay 2000; maxDelay 0 total Acqs 100001; Acqs without change 0; fail metaMti 100000; metaDti 0; metaMAT 0; metaMUT 0; metaDUT 0 measureTIrangeAndOffset() failed ET cal failure (SLOW FAST) value = 0 = 0x0 rtlStartup value = 0 = 0x0 tsComp value = 0 = 0x0 ################### topScript ends ############################### Done executing startup script c:/vxboot/topscript.hw KERNEL: RUNNING TIME SLICED AT 10 TICKS PER SLICE! -> DIAGNOSTIC EXECUTION begin minMain 2000; maxMain 0; minDelay 2000; maxDelay 0 total Acqs 100001; Acqs without change 0; fail metaMti 100000; metaDti 0; metaMAT 0; metaMUT 0; metaDUT 0 measureTIrangeAndOffset() failed minMain 2000; maxMain 0; minDelay 2000; maxDelay 0 total Acqs 100001; Acqs without change 0; fail metaMti 100000; metaDti 0; metaMAT 0; metaMUT 0; metaDUT 0 measureTIrangeAndOffset() failed ET cal failure (SLOW FAST) OptionST not tested. (option not enabled) BK2 not tested (option not enabled) DIAGNOSTIC EXECUTION finished Started SPC ( Calibration ) minMain 2000; maxMain 0; minDelay 2000; maxDelay 0 total Acqs 100001; Acqs without change 0; fail metaMti 100000; metaDti 0; metaMAT 0; metaMUT 0; metaDUT 0 measureTIrangeAndOffset() failed minMain 2000; maxMain 0; minDelay 2000; maxDelay 0 total Acqs 100001; Acqs without change 0; fail metaMti 100000; metaDti 0; metaMAT 0; metaMUT 0; metaDUT 0 measureTIrangeAndOffset() failed ET cal failure (SLOW FAST) minMain 2000; maxMain 0; minDelay 2000; maxDelay 0 total Acqs 100001; Acqs without change 0; fail metaMti 100000; metaDti 0; metaMAT 0; metaMUT 0; metaDUT 0 measureTIrangeAndOffset() failed minMain 2000; maxMain 0; minDelay 2000; maxDelay 0 total Acqs 100001; Acqs without change 0; fail metaMti 100000; metaDti 0; metaMAT 0; metaMUT 0; metaDUT 0 measureTIrangeAndOffset() failed ET cal failure (SLOW FAST) --- End code --- |
| Guy_Ellis:
Hi Charlyd, I have a second TDS7404 here that looks promising. It had a faulty NVRAM chip which I have replaced, and fitted a new battery. I have initialised the NVRAM, and it boots. I haven't had a chance to run SPC yet, but I have attached the logs below for you to compare. I've had a look at your logs and wonder if it's is a good idea to use cal files from another scope? I'm pretty sure these can be regenerated - we just need to figure out how to do that? Regards, - Guy. CPU: Tektronix HPP - PowerPC 740 Version: 5.3.1 BSP version: 1.0.8.0 Creation date: Jun 27 2003, 09:52:32 Press any key to stop auto-boot... 0 auto-booting... boot device : sm processor number : 0 file name : c:/vxboot/vxWorks.st inet on backplane (b): 192.168.0.1:ffffff00 host inet (h) : 192.168.0.2 gateway inet (g) : 192.168.0.1 flags (f) : 0x1008 target name (tn) : tds7000 startup script (s) : c:/vxboot/topscript.hw Initializing backplane net with anchor at 0x1e80000... done. Backplane anchor at 0x1e80000... Attaching network interface sm0... done done. Attaching network interface lo0... done. PC1 dio device driver installed Attaching DIO client... done. Loading c:/vxboot/vxWorks.st...1773516 Starting at 0x200000... Host Name: bootHost User: target Initializing backplane net with anchor at 0x16e0000... done. Backplane anchor at 0x16e0000... Attaching network interface sm0... done done. PC1 dio device driver installed Attaching network interface lo0... done. Adding 6008 symbols for standalone. ]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]] ]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]] ]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]]] ]]]]]]]]]]] ]]]] ]]]]]]]]]] ]] ]]]] (R) ] ]]]]]]]]] ]]]]]] ]]]]]]]] ]] ]]]] ]] ]]]]]]] ]]]]]]]] ]]]]]] ] ]] ]]]] ]]] ]]]]] ] ]]] ] ]]]] ]]] ]]]]]]]]] ]]]] ]] ]]]] ]] ]]]]] ]]]] ]]] ]] ] ]]] ]] ]]]]] ]]]]]] ]] ]]]]]]] ]]]] ]] ]]]] ]]]]] ] ]]]] ]]]]] ]]]]]]]] ]]]] ]] ]]]] ]]]]]]] ]]]] ]]]]]] ]]]]] ]]]]]] ] ]]]]] ]]]] ]] ]]]] ]]]]]]]] ]]]] ]]]]]]] ]]]]] ] ]]]]]] ] ]]] ]]]] ]] ]]]] ]]]] ]]]] ]]]] ]]]]]]]] ]]]]] ]]] ]]]]]]] ] ]]]]]]] ]]]] ]]]] ]]]] ]]]]] ]]]]]]]]]]]]]]]]]]]]]]]]]]]]]] ]]]]]]]]]]]]]]]]]]]]]]]]]]]]] Development System ]]]]]]]]]]]]]]]]]]]]]]]]]]]] ]]]]]]]]]]]]]]]]]]]]]]]]]]] VxWorks version 5.3.1 ]]]]]]]]]]]]]]]]]]]]]]]]]] KERNEL: WIND version 2.5 ]]]]]]]]]]]]]]]]]]]]]]]]] Copyright Wind River Systems, Inc., 1984-1997 CPU: Tektronix HPP - PowerPC 740. Processor #0. Memory Size: 0x2000000. BSP version 1.0.8.0 . WDB: Ready. KERNEL: THERMAL CONTROL INITIATED Executing startup script c:/vxboot/topscript.hw ... dbFsIoctl: invalid code 3 ################### topScript starts ############################### cd "dbfs:c:/vxBoot" value = 0 = 0x0 ld < TekScope.o value = 23666688 = 0x1692000 ld < TekScopeSM.o value = 23666048 = 0x1691d80 pDbDev = ramDevCreate(0, 512, 1152, 1152, 0) new symbol "pDbDev" added to symbol table. pDbDev = 0x1693940: value = 23673184 = 0x1693960 = pDbDev + 0x20 pDBfs = dosFsMkfs("/db", pDbDev) new symbol "pDBfs" added to symbol table. pDBfs = 0x1691bc0: value = 23665696 = 0x1691c20 = pDBfs + 0x60 copy ("TekScope.DBF", "/db/TekScope.DBF") value = 0 = 0x0 overrideFPdefaults() value = 0 = 0x0 measurePPCtimebase value = 0 = 0x0 registerMemPartitionSizes ( 0x00000000 ) value = 0 = 0x0 semInitForQueueCheckLogMsg value = 21818112 = 0x14ceb00 initSysWaiter() value = 0 = 0x0 raxInit value = 0 = 0x0 paiInit done value = 16983892 = 0x1032754 = PaiBaseRegisterTable + 0x948 verifyAcqBoardID value = 0 = 0x0 tcsLoadTableToNVRAM "pc:c:/vxBoot/Setup.tcs" value = 0 = 0x0 updateAcqProcFirmware("dbfs:c:/vxBoot/acqProc.hex") 0x16987e0 (tShell): AVR FLASH contents match file dbfs:c:/vxBoot/acqProc.hex - no programming needed. value = 1 = 0x1 cd "pc:" value = 0 = 0x0 raxInitializeDevices done 0x16987e0 (tShell): (Re)starting Acq Processor value = 21815208 = 0x14cdfa8 excHookAdd(stopStressForException) value = 16770764 = 0xffe6cc = stopStressForException dsyInit value = 21806440 = 0x14cbd68 dsyWPApplyNextRegs value = 16973768 = 0x102ffc8 = DsyRegisterTable + 0xab0 # remove '#' from next line to set powerup scope state to factory default: # NvramClearDb = 1 nvram_initFileSys value = 21801184 = 0x14ca8e0 TmgrInit("/db") value = 0 = 0x0 PhxProxyInit value = 20560480 = 0x139ba60 rtlInit value = 0 = 0x0 sysExtMemAddToPool memSysPartId,0x02000000,0x0e000000 value = 0 = 0x0 hwAccount_init sysEnetAddrGet: 9d:0c:f2:b3:66:9f value = 0 = 0x0 GpibInit value = 267500128 = 0xff1ba60 wfmMemInit(1) value = 57781984 = 0x371aee0 vertInstall value = 57774948 = 0x3719364 calConStatus = cc_init(2) 0x16987e0 (tShell): Reading 300 bytes of service cal data from EEPROM[0] into 0x3719220 0x16987e0 (tShell): nvMem_read succeeded 0x16987e0 (tShell): Reading 82 bytes of service cal data from EEPROM[300] into 0x3719300 0x16987e0 (tShell): nvMem_read succeeded 0x16987e0 (tShell): cc_restoreBank: can't open file C:/vxBoot/calCons/0 0x16987e0 (tShell): cc_init: bank 0 chksum failed, NOT reset to factory 0x16987e0 (tShell): cc_restoreBank: can't open file C:/vxBoot/calCons/1 0x16987e0 (tShell): cc_init: bank 1 chksum failed, NOT reset to factory 0x16987e0 (tShell): cc_restoreBank: can't open file C:/vxBoot/calCons/2 0x16987e0 (tShell): cc_init: bank 2 chksum failed, NOT reset to factory 0x16987e0 (tShell): cc_restoreBank: can't open file C:/vxBoot/calCons/3 0x16987e0 (tShell): cc_init: bank 3 chksum failed, NOT reset to factory calConSta0xtus16987e0 (tShell): = cc_restoreBank: can't open file 0xC:/vxBoot/calCons/13535528 0x16987e0 (tShell: value = ): cc_init: bank 332635 chksum failed, NOT reset to factory = 0x0x81ef16987e0 (tShell): cc_restoreBank: can't open file C:/vxBoot/calCons/ 6 vertPostCalRest0xoreInstall16987e0 ( tShell): cc_init: bank value = 6 chksum failed, NOT reset to factory 00x16987e0 = 0x (tShell): cc_restoreBank: can't open file 0C:/vxBoot/calCons/ 7 0x16987e0 ( tShell): cc_init: bank MathMeasCr7eateTheMathandMeasu chksum failed, NOT reset to factory rem0xentLibraries16987e0 (tShell): cc_restoreBank: can't open file value = C:/vxBoot/calCons/576023688 = 0x0x16987e0 (tShell36ef140): cc_init: bank 8 chksum failed, NOT reset to factory initAcqMap0xper16987e0 (tShell): cc_restoreBank: can't open file valuC:/vxBoot/calCons/e 15= 0x16987e0 (tShell0): cc_init: bank 15 = 0x chksum failed, NOT reset to factory 00x16987e0 (tShell ): Optical hardware is not present. 0x trigMapInit16987e0 (tShell): initAcqMapper: Enter value = 0 = 0x0 dsyMapInit value = 0 = 0x0 mathMapperInit value = 0 = 0x0 utilMapInit value = 57599712 = 0x36ee6e0 TmgrHookup value = 0 = 0x0 PhxProxyHookup value = 0 = 0x0 hookupAcqMapper value = 20074648 = 0x1325098 = AcqMapperRtlDataList MathMeasMapHookup value = 23692552 = 0x1698508 dsyMapHookup value = 23692656 = 0x1698570 utilMapHookup value = 23692688 = 0x1698590 rtlHookup value = 0 = 0x0 horizMapHookup value = 23692648 = 0x1698568 fastFrameMapHookup value = 23692672 = 0x1698580 trigMapHookup value = 23692680 = 0x1698588 vertMapHookup value = 23692672 = 0x1698580 spcMapHookup value = 23692672 = 0x1698580 autosetMapHookup value = 23692712 = 0x16985a8 timeSyncMapHookup value = 23692704 = 0x16985a0 auxTrigMapHookup value = 23692688 = 0x1698590 dpoMapHookup value = 23692680 = 0x1698588 TmgrStartup value = 56638608 = 0x3603c90 PhxProxyStartup value = 56588064 = 0x35f7720 taskDelay(2) value = 0 = 0x0 gpibHwInitC value = 0 = 0x0 gpibTaskSpawn value = 56365216 = 0x35c10a0 gpibIoInit "/dbios/DBIO005","/dbios/DBIO006" value = 55878048 = 0x354a1a0 enablePowerupDiags = 0 enablePowerupDiags = 0x12fad08: value = 0 = 0x0 EnableHorizInterpolation=1 new symbol "EnableHorizInterpolation" added to symbol table. EnableHorizInterpolation = 0x365b480: value = 1 = 0x1 runPowerupDiags done dipSwitch = 0x0, enablePowerupDiags = 0, puMode=0 CalDiagPowerUpDiagsEnabled = 0 !! SKIPPING POWERUP DIAGNOSTICS !! value = 0 = 0x0 rtlStartup value = 0 = 0x0 tsComp value = 0 = 0x0 ################### topScript ends ############################### Done executing startup script c:/vxboot/topscript.hw KERNEL: RUNNING TIME SLICED AT 10 TICKS PER SLICE! -> |
| charlyd:
Hi Guy, Yes this for sure sound good, to figure out how we can get the right CalCons files in place. |
| Navigation |
| Message Index |
| Next page |
| Previous page |