I am aware that this is an old thread, but searching on Duckgo for Tektronix 134 repair this is the link that makes the most sense, and probably the only source of information about someone actually having repaired their 134 unit.
Purchased my type 134 amplifier a couple of months back when I saw one going for around $25 on ebay (including original power supply) in anticipation that I would eventually get a matching current probe at some point and at a decent price. There are some crazy prices out there for these type of probes. Also got the hanger kit and the original BNC-F to BNC-M jumper wire both still sealed to complete the set.
So last week finally managed to get a P6021 current probe and tested it with the 134 on my scope by setting it to AC 50mV/div as indicated, to measure the current of about 1.7V 1KHz signal from my audio generator into a 600 ohm (actually 617 ohms) 1/4 watt resistor. The measurements where very close to the expected, perhaps just a tiny bit on the high side.
Verified the signal amplitude from the generator on my HP974A DMM, which indicated 1.71V( rms), so 1.71 / 617 = 2.77mA then converted to Pk-Pk to match the scope reading would be x2.828 = 7.83mA P-P. On the scope I see 4 divisions at 2mA/div scale so just about 8mA. Slightly above the expected reading but still very close.
So far so good, but after reading the 134 instructions I saw that the scope has actually to be set to DC coupling 50mV/div, but when I did that the trace went way off scale. And the reason is that I am getting a constant 5.8V DC output from the 134 amplifier at all times, with or without any signal input.
I checked the schematic and there is C165 a ceramic capacitor of 1uF 25V just between the amplifier output and the BNC output connector, which should completely block any DC voltage. I opened the 134 and everything inside looks undisturbed and original, so I'm guessing C165 must have gotten bad (leaky), but I wonder how this could happen to a ceramic cap. The only thing might be if someone connected a voltage higher than the 25V rating of the capacitor to the output of the 134. I don't think ceramic caps ever go bad just because old age. They are rather very stable.
Anyway, will start doing further checks later this evening. Any comments or suggestion welcome.