I tried a few more component extractions for testing that had been mentioned in other threads. The Q935 thyristor, C925, C943, and R898-R891 - no luck there. So I went for the addition of a 43V PSU across the two test points. The first PSU I tried didn't work so well and I switched it off as I thought things were getting hot. Second one I tried caused the classic tick, tick, tick and two flashing neons of a resetting power supply. Carefully touching the board for hot components, and visual inspection, showed a bit of shadow on the PCB around CR907. Sure enough the smoking diode - a dead short.
It is marked as UES1106, which is a fast recovery diode, not something I keep in stock. Not sure what part it has to play in the circuit when bypassing the original supply, but with it removed, and also Q935 and Q9070 out of circuit, I get the power LED on and no ticking

. I once briefly saw a fuzzy rectangle on the display when I pressed BEAM FIND

, but it has stopped doing that now

. The PSU is drawing about 700mA in this state. Do I need to replace that diode, or is it no longer part of the equation with the new 43V PSU in circuit?
Getting closer.