Yeah, I bet it is a measurement error. I have checked r935 - spot on 22,1k.
C935 is not ceramic cap. It is a film capacitor in cubic plastic box. I connected 22n over C935. Frequency went down to approx 560Hz but no visible change on display in alt mode, nor in chop. I was connecting, and dissconnecting it live to see display in realtime. Main problem is somewhere else I think.
I will extract this cap tomorrow from board and check capacitance. If out of specs - I will change it. Also will change R935 to 24K if i have in stock, or add 2k resistor in series.
I checked in circuit also: R934=1k, R938= 2,2k, R939=1,2k, but 4,1V is 4,09 so... I will test all calibrator circuit parts out of circuit tomorrow.
In theory, one could temporarily disable U930B by grounding pin 6, then inject a 1 kHz square wave of suitable amplitude into the CLK 1k point, to see if the chop anomaly disappears.
Also will check that, but considering above, there propably will be no change on display.
Hmm. I forgot to mention before but I was playing around with service menu. If I go to Proc board menu -> Readout menu -> Exercise RO interfece, Flicker is much more noticable. From theory of operation This forces communication between Measurement processor and RO processor. It's AD0-7, then D0-7, then R0-7. So maybe something that is connected to those lines is playing with us. Sorry I don't have enough knowledge and experience to troubleshoot this myself. Maybe I am totally wrong.