Hi all,
I am new member to this site,
Please can you help,
I purchased a Tek 2440 scope on ebay some years back the fault listed by the seller was noisy trace but all self tests passed. anyways last year during lock down I took a look inside and found the Rifa X class caps bulged on the power supply so I changed them for the same make and rating ( at the time I had not realized the Rifa Madness as Dave would say) so after installing the scope was working 30 mins later Snap Crackle Pop a lot of smoke, so killed power and waited for the smoke to clear, I open her up and found the Rifa C816 blown out. lesson to learn should have found Dave's video on Rifa madness before replacing!
So I'm into second lock down here in Guernsey thought I would take another look. so I am going to change the line Rifa caps with a good replacement but the resistors in series are burned, The R809 is simple as its a 30 ohm 5% .5 watt, but the resistor next to C706 (RT805) seems to be a thermal resistor of 5 ohm 10% Part number 5DA5R0K-270 SIL or Tek P.N. 307-0157-00 but I cant seem to find any thermal resistors that are less the 1K? let alone OEM parts!
So questions are as follows;
Can anyone shed some light as to what alternatives can be placed in this position (RT805) and do I need to worry about it as I am on 240 Volt line not 120 volt?
Subject to the above having a successful work around. Can anyone advise on where I need to look for power supply ripple that could be causing trace noise? ( I have an old school Iwatsu 40 mhz scope that I can use to check)
I plan to change out the Dallas NVRAM chips with new and install sockets when the time comes, but as they are working and the scope passes self test, I am thinking best left alone for now.
Many thanks all and sorry if this has been covered already.
Kind regards
Denzil