I was wondering if anyone could help me out with an issue I'm having.
Perhaps it has been seen before??
My 2445A starts running fine. The display and everything is okay.
After a short period of running between 10-20 minutes the screen will start to become jittery and then suddenly the display will disappear with a faint glow on the left side of the screen. If you adjust the horizontal position of the waveform to be as far right as possible sometimes the end of the waveform will just appear on the left hand side of the display.
Sometimes this issue can be resolved by pressing the "beam find" which will display a perfectly centered small version of the image. The image will then return to a normal or jittery state in the center of the screen.
More often though, "beam find" will display a perfectly centered small version of the image but after releasing the beam find button the image will return to the left side of the screen.
Update:
I have checked all the voltages on J119 and all of them check out as okay both before and while the error is occuring.
I have done some reading through the service manual found here:
https://www.tek.com/manual/2445a-and-2455a-service-manualI've been trying to figure out what the beam find button does in the circuits operation but I can't find anything to explain it. I'm doing this because beam find seems to be able to recenter the image without issue and I'm hoping that it might point me in the right direction for a fix.
I'm afraid that this may be an issue with U800 which is a custom chip or an issue with the high voltage supply for the CRT.
Has anyone any experience with such issues or could point me in the direction of where to check?
I have fairly good experience with electronics so I believe I can complete a repair I just need help trying to track down the source of the problem.