Author Topic: Tektronix 2445a  (Read 1018 times)
Offline kaborexTopic starter
Tektronix 2445a
« on: December 01, 2017, 01:29:05 pm »
Hi all,
I am new here.  interested in buying a scope to fix.  I have seen one with the following issues, any ideas where to start troubleshooting or what could be causing this?
The unit comes on initially and works fine approx. 5 minutes, after which time the x scan collapses.
regards
Kaborex
Offline tggzzz
Re: Tektronix 2445a
« Reply #1 on: December 01, 2017, 04:34:38 pm »
Presumably you are too late; this
https://www.ebay.co.uk/itm/Tektronix-oscilloscope-model-2445A/152800788852
has already been sold.
Apart from that, there are many sources of information on this forum and at the Yahoo! Tekscopes group.
Offline David Hess
  • DavidH
Re: Tektronix 2445a
« Reply #2 on: December 01, 2017, 10:24:59 pm »
I recommend avoiding that one.  While this is a known problem with the 24xx series analog oscilloscopes, repair requires a replacement hybrid from a parts mule.
If you are looking for an oscilloscope to repair, then I recommend something older which will be easier to diagnose and maintain.