Hello
Firstly, I did search but could not find an equivalent problem.
I have a Tektronix Oscilloscope 2445B Serial # B063339
(the baby brother of the 2465B)
History:-I replaced four leaking capacitors on A5 Control board over 2 years ago,
all tracks around the area were cleaned and found to be intact and
connecting fine with everything working perfect since then.
Now:-Powers up OK, everything looks normal.
Run self tests = ALL 00 PASS
All Power Supply Voltages and ripples at J119 check out OK.
Errors:-Control Board A5...
The -1.25V reads -1.01V
The +1.36V reads +1.103
Using the ''Vertical Position'' controls for any channel it's impossible to set the trace in the centre of the screen.
(however with an input signal applied the Vertical amplifiers seem to be working as normal)
Horizontal system will not trigger to any input signal.
CH1 or 2 with no input signal, if I select
50Ω DC (50 Ohm DC) the
50Ω OVERLOAD error appears at t he top of the crt,
relays click and the input resets to GND, if I up select either to DC GND or AC the
50Ω OVERLOAD error MSG disappears.
I've looked through the manual but I can't seem to work out what circuit sets the -1.25V and +1.36V references.
I would appreciate any help or advice on the problem or what to check.
Thanks in advance for looking