I'm opening a new thread here because I need to fix something here.
I already wrote in another thread. Here again briefly the summary:
I purchased a defective Tektronix 2457B with Option 9 and 10. Error was, that the device turned on and off again and again. Even the smell showed that electrolyte must have leaked out. The device stank like a trash can. After I opened the device, I discovered the leaked SMD electrolytic capacitors on the A5 controller board. The traces were not attacked and also neighboring components were still OK. I checked that under the microscope. After I had cleaned everything and used new Elkos, the device worked again. The error did not occur anymore. Otherwise, I could not detect any abnormalities in the power supply, all values are OK, Ripple unobtrusive.
The flicker (looks like modulation of the z-Axis) occurs only in connection with the ARB function generator and only in the range between 1-10Mhz. The generator does not do more. The signal representation is otherwise first class.
The flicker is independent of the position of the readout knob. But others also wrote that, who have this flicker too. It has nothing to do with the readout, it also occurs when the readout is turned off. It has a very slow frequency, anything around the 2-3 Hz, it looks like waves of a modulation. CRT is set with Cal 08.
In the meantime, I have noticed that the flicker occurs at frequencies of 1MHZ and more, if the value is exactly 1MHZ, 2MHZ, 3MHZ, etc., that is set exactly to 7 digits on the function generator. So for example 5,000,000.0MHZ. Therefore, this does not occur in practice and also with analog signals. If I adjust the value by only 10Hz, the flicker disappears. So at, for example, 5,000,010.0MHZ the flicker disappears. So there seems to be something to do with driving the Z-axis, any interference.
Does anyone have any idea where to start with the search? I think that is a defect, otherwise it would occur more often.