Some more news on the meter.
First of all let me show you some pictures.
The outside: Aluminium enclosure, trendy brown colour and very deep. Relatively clean and in general in good condition.
Now with the hood up.
On the lower right side you can just about see the +/- 12V regulators, both of wich were dead.
To the left of them there's the whole bank of relays for the input selection/attenuation. The right-most relay on the piece of perfboard is a bodge that I made. There used to sit a reed relay like the others around it which was shot. It's a DPST type and obviously I didn't have one on hand so made that adapter. More on that later. Just off the centre to the left are two relays that also sit on a little adpter board. I have no idea why or if it is original. There are three identical footprints on the main board, one is populated by the proper relay for the footprint the other two sit on that board. Repair maybe? Beneath this adapter board lies the attenuator resistor network, a FN207 made by a company called A-B.
The cards from top to bottom.
- The 8085 controller board, ribbon cables go off to the frontpanel.
- What I'd call the main A/D converter board, A/D conversion is done by LD120/LD121.
- The TRMS converter board, apparently an option not fitted to all of these meters.
- Not quite sure about that one, probably the resistance range, might do more as it seem a bit too complex for only measuring resistance.
At the moment I'm a bit confused by the unit. There is evidence of heavy modifications and I dont know who did it or why they did it. I'm pretty sure it's been done by one person or organisation, might even have been done by Siemens themselves.
There's the thing with relays I mentioned above but there's more. On the mainboard there are a few bodge wires and even what looks like a couple of bodged in parts. There are a few of those on the plug in boards as well. It looks like somebody's drilled the PCB's and added proper solder pads for those bodged in parts. I have no idea if that is possible, but it shure looks like it.
Here on the AD card you see the little blue trimpot in the top left corner. It uses one of the fiducials and two holes with proper pads around them. The three connections to the rest of that board are made using wires. The pot sits on top of a QC sticker.
Then there's the relay on the Ohms board, you can clearly see the footprint of the original part and what is in there. It's got proper solder pads, too. Not sure whats going on there
.
The wires used to wire in all the extra components and bodge connections are all off the same type, so chances are it was all done by the same person.
The only visible damage I could find so far was this blown track (centre of image): There was some scorching around the relay contact next to the solder lug. Look at the PCB next to the solder lugs, the trace ran from pin 8 of K8 to the via on the right of the two solder lugs. It looks like there was a flash over from the relay to this solder lug, the jumper wire between the lugs lets you disconnect the -12V rail.
The missing piece of trace was vapourized. These reed relays are used for the mA ranges, so I suspect someone exceeded the max. allowed voltage, that created the flash over and killed the PSU.
That's the back of the mainboard right under the supply section. You might be able to see some of the double cut tracks I had to repair. I've not taken a pic of the before state.
These same green wires are used on all the mods in this unit, they were glued to the PCB with hot glue once. Most of the hot snot has come off the board over the years.
I've now replaced the 4000 series IC's on the ADC board, they were all blown but that hasn't really changed anything. I still get random numbers on the display. There is some reaction to an applied voltage on the inputs but not really predictable.
If somebody has any idea how to check if the LD120/121 are fried please do leave it below, anyway I'm expecting a quote for a set of them.
The PSU looks fine now, I'm confident that problem is solved.
The manual is on its way, can't really continue doing anything sensible without that.