I got this beautiful 1960s scope used, saw a green dot or something few seconds months ago, but I didn't want to risk it with its 1960s electro. cap's, so I stopped until caps got here.
I just recapped the LV PSU, which is modular and separates fully off the back of the scope. So it's just running its bleed resistors. So checking some voltages, the +12V reg might be bad. And the high voltage astable multi-vibrator is not working.
the full schematic is on page 35
http://w140.com/tek_422_ac_powersupply.pdfI'm running off 120VAC, the transformer is wired right and the outputs from it, and on the other side of the rectifier diodes, seems about right (if a little high since it's supposed to be 110V or 115VAC)
In the 55V reg at the top of R655, which is 5.6k and ok, I get 79.1VDC (instead of 63V from 110VAC but I think it's ok)
The bottom of R655 is 38.8V (schematic says 34.5V, IDK if thats with the scope running tho or not) But thats probably fine.
Reliant on the 55V reg is the 12V reg
Q653
Vc=26.3V
Vb=2.9V should be 13.2V how can that be fine, or is it just with no real load ?
Ve=2.9 should be 12.5
R654 6.8K is good, Q653 reads fine out of circuit(they plug in and I have continuity on the plug)
The +12v output from q657 Ve is 18.6, but with just a bleed resistor, so maybe thats ok ?
The HV driver is not working right
HV driver
q675 (c/b/e) -14.04 -14.2 -14.5
q685 -14.1 -14.2 -14.5
It shows the collector waveform in the schem, but all I see is it going from -10 to -15V, at 120Hz, and a lot more like a 30degree triangle < than a Sq wave pulse generator.

? I'm pretty new at this, both those BJTs read fine in-circuit, so what's going on

Or what do I not know/realize ? I've measured /checked most things and they seem fine too