I suppose you have already tried the focus adjustment.
Here is a quote from another group that may be pertinent, even though it is about another Tek Scope:
I'd run the scope by an ex-Tek CRT inspector, and she thought the symptoms
looked like "low cathode". I interpreted this to mean that, although emission
is still good, it is coming predominantly from parts of the cathode structure
that are not optimum for a nicely focused beam. Another CRT related
possibility mentioned at the time was some kind of mechanical damage that affected the
internal alignment of the elements. Whatever the root cause, it turned out the
problem was indeed with the CRT itself, and not the electronics. A new CRT
(actually for a 465M) from ebay, a new blue filter from Walter at Sphere
Research, and a recal, and it seems to be as good as rain again. With the new tube,
the traces are nicely crisp. Not as good as I remember a 485 mind you, but
quite nice, and certainly not worse than any other common Tek scope of the 80's
and 90's that I'm familiar with.