Hi all,
was just about to open a new thread when I came across this thread, which is a very similar problem to mine, so perhaps we can help others in the future by keeping a
'low -8v on a 465' in one thread?
Last night I was taking pictures of my lovely old 465 as I intend to sell it, and took some pictures of perfect square waves on the CRT, then just left it running whilst doing other stuff.
When I next checked - perhaps an hour later - the trace had disappeared!
Upon further inspection and with the Intensity increased, however, there is still one small dot - with, or without the beam-finder.
As the case was already off, a thorough search for burnt or blown caps or components revealed nothing out of the ordinary. What was out of the ordinary was TP1568, the minus 8v Test Point, which had been minus 7.9v or so and now reads minus 4.89v.
This is why I am here...
Reading through earlier posts in this thread, there was a page linked with the procedures for isolating the low-voltage power supplies (see attachment), which says to unsodler (it does

) one end of various components, and this is where I became confused. One of the components is R1445...
R1445 is a three pin trimpot!
(*Edit... just realised that the linked page is for scopes above S/N B250000 - Mine is 100123)
Furthermore, using continuity on a DMM there seems to be no connection between TP1568 and any of the components listed in that page of the manual, so how would that help?
No component seems to be overheating although two PNP's Q1262 and Q1282 are too hot to touch, but test totally fine.
I have spent all day reading different versions of the 465 SM, without finding any version with that particular page, nor clear instructions as to how to progress, so am having to put my hands up, as short of removing and testing every single component I can identify in the -8v line, I don't know where to test.
Someone must have some CLEAR idea of how to resolve this problem.
Many thanks in advance.