Your picture makes the trace seem very bright.
Have you tried and failed to calibrate the existing probes? Be aware the only purpose of the cal out signal is to allow the edge to be flattened out when seen with a standard *10 probe. The amplitude and frequency are not well defined. The probe's "capacitance compensation range, e.g. 10-30pF, must encompass the scope's input capacitance, e.g. 20pF.
For all we know, you may have the controls in the wrong position for the B channel to trigger.