Hello, this is my first post on the EEVBlog so I've been dragging my feet until I finished reading a bit more of the Tektronix manual. While also exploring other repair posts that might help with my scopes fault. I'm a bit of a novice with more complex electronics even though I took a few electronic classes in collage. I even built a basic Theremin from scratch as a very successful birthday gift. So I would say I know enough to stay safe and double check myself before testing anything in this scope. But at best I'm still just a self taught hobbyist.
I bought this scope on ebay as not working with no trace. After winning the offer I started to think that it was a mistake and I was now the owner of a boat anchor. Thankfully in less then 5 minutes of tinkering I was able to get a trace. This Tektronix 465B was eager to show it's stuff and that was a huge relief. Now starts the fun part, cleaning and fixing this Tek.
Besides some Dirty contacts the only other issue I can see is bad focus. Also the Intensity seems a little too touchy to me.
Looking in the manual I can see there are some test points and wave forms I can check... Under the high voltage cover plate. Oh boy.
It's a bit dirty under there and from what I read in Tek Troubleshooting PDF That dust can cause leakage. So I will be cleaning that section... very carefully.
http://www.sphere.bc.ca/test/tek-parts/troubleshooting-scopes.pdfI'm unsure that the dust could cause the issue though. I hooked up my working Leader scope to Test point 93 shown in the CRT Circuit schematic and the waveform looks ok. Until I adjust the intensity and this happens.
I was a bit nervous hooking up my scope because the test point is under the high voltage caps. So I unplugged the scope, connected the probe one handed, and then plugged it back in and turned it on. Being extra careful to use only one hand and keep away from the high voltage.
So at the moment I'm a bit stuck. Did I measure test point 93 correctly? Is it possible that my probe is giving this weird signal? Part of the signal looks like the waveform in the manual but why the unchanging signal when the intensity is turned?
Knowing that both of my cheap multimeters are only to be used as paper weights, I went on the hunt for a DMM with better specs than what the manual calls for. Having found a Keithley 178 I checked the voltage rails and they seem to be in spec with the manual. Though I'm sure my Keithley should be calibrated, it's still a whole lot better then those cheap multimeters.
I tested the -2450 volts with a working triplett 630 NA. It was spot on. The manual recommended the Triplett 630-NA so I gave in and bought one. I was very careful testing the high voltage following the Triplett manual. Hooked it up with the scope unplugged, then plugged the scope back in and with one hand turned the scope on. I must admit I was nervous about using this old meter but rather pleased that it worked flawlessly. Also I did check the meter to make sure it was clean both inside and out before even thinking about using it to test the high voltage. I did not want to damage my new favorite tool.
Also I was able to test the Keithley against the Triplett and was rather surprised they complemented each other. Yes it's analog against digital, but still I rather like my new tools. They instill more confidence then any cheep multimeter out there.
The CRT Bias was checked/set to the adjustment procedure outlined in the manual. It seems to work as it should and measures fine with a scope hooked up to TP4217. Although maybe an experienced eye can spot something in the trace... hmm ok, I just checked it and this is something I didn't notice before. While adjusting the intensity I got this at TP4217. That can't be good, and this changes where I should look in the schematic. Bad caps I hope?