Author Topic: Tektronix 475 No Trace  (Read 3066 times)
Offline michelk005@gmail.comTopic starter
« on: April 29, 2015, 12:54:27 pm »
I have a 30 year old Tektronix 475 that I have a problem with.  It powers on, no trace.  I have opened it up and per the service manual checked all the LV power supply values and they are all within specs.  What I believe is the issue is no High Voltage.  Does anyone have any experience with a similar problem that was related to the High Voltage section that can share any troubleshooting thoughts.
Thanks,
Keith
Re: Tektronix 475 No Trace
« Reply #1 on: April 29, 2015, 01:02:13 pm »
Offline Addicted2AnalogTek
« Reply #2 on: May 05, 2015, 03:41:47 pm »
I've been starting my old Tek scope repairs by doing a very thorough cleaning of all the switch contacts and dripping some deoxit down the pot shafts and working them around, and also re-seating all of the socketed components.  If it hasn't been used/running properly fairly recently, then this would be a great place to start.  Then you can get a good idea of what is really wrong without having to worry about bad contacts.
On both my 465 and 468, doing this solved a number of issues that the scopes had.    Hope this helps!