Hello together,
seems like I got lucky and got my first scope straight from the dumpster. Sadly that poor Tek 475 seems to be in need of repair

. It powers up fine and I can see the beam in the center of the screen.

But here the problems begin. If I want to move the vertical position of channel 1 it does not behave like I would expect it. Instead of moving the dot vertically it moves diagonally across the screen from bottom left to the top right.
When I attach a probe to the calibrator output the same happens. Now I see two dots which I expect to be the peaks of the square wave outputs. Changing the Volts/div works as I would expect it. But changing anything in the timebase does nothing as well as changing anything in the triggering section.
Does anybody here have experience with those old scopes and can point me into the right direction? From seaching the forums I already found out that the main caps on these scopes tend to fail. I measured all voltages and they are spot on. With my cheapie Aneng multimeter I also see no ripple in AC Volts mode (except a couple mV). Sadly I do not have another scope to check the ripple
Thanks for the help
Markus