Last I bought a whole bunch of partly broken devices. Included was a 475 with DM43. CH1 was dead and the high-voltage eletrolyte cups shown a ripple of about 40VPP. Have then exchanged all Elkos, even if only 2 were really broken.
Ch1 did not work because the vertical channel switch did not turn the signal thru. First thought the hybrid would be broken, until I noticed that there was an error in the logic control. One Resistance was high impedance. After replacing this, CH1 alives. The DM43 showed no plausible values, only 3.3.3.3: in the power supply of the device (hidden under a cover I found a power supply), in which a 12V voltage regulator was defective. A pain to expand the part.
After exchanging some tantalum for Low-ESR electrolytic capacitors (with 100n SMD ceramic between the legs) and normal tantalum, the device works very well. It triggers extremely clean, (much better than my Rigol) the beam is clean and sharp. I'm really excited about the device. Also included was the appropriate Scope mobile.
A 2467B was also there. Is already repaired.