yes I understand .. reference oscillator is warm, I can't find out more yet
Ok, what I have done so far ..
1. I read user manual

2. I read Service manual Vol1, and Vol2., not all yet, it is too extensive
3. I read the links about repair this SA,
https://iz3kgj.jimdofree.com/realizzazioni/tektronix-495p-phaselock-failure-1st-lo-repair/ Ind more useful service notes here ..
http://www.ke5fx.com/49x_notes.pdf and all repairs have already been made.
4. I did a check Aids from diagnostic menu - all Expected and measured Frequency and ofsets (1ST Lo, 2nd Lo, DAC ) are in tolerance.
5. All PSU Voltages correspondent with Service manual, Power status LED is Green.
6. I repaired Intermitted contact on 2nd LO, (SMA Connector)
7. Visual inspected Whole boards for leak, corrosion etc..
Result ..
Basic calibration not working, Calibrator SIGNAL NOT PRESENT ..
INT OSC shows I=U, sometimes when i change Center Frequency with knob for a moment shows INT
When I change SPAN under 100kHz sometimes 1ST Lo , 2nd Lo error appears ..
I will need more equipment to continue ..