Tektronix 576 CRT Question
« on: March 06, 2023, 09:31:45 pm »
I recently bought another 576 this time with blurry CRT dot and trace.
Wondering if any of you CRT experts could help me with a question?

Checked all voltages as per manual and all are within spec.
My problem is the trace... I have to turn the Intensity knob 3/4 way up to get a dot and then I can focus but it only lasts for a few moments and goes out of focus again.

I have worked the Invert, Zero and Cal buttons a number of times and I do get a more focused trace for a minute or so and then back to being unfocused.

What I am wondering, is this a sign of dying CRT or are there resistors and Capacitors that need to be checked.
As I said, I have 4000 volts on CRT and all other voltages are within spec. I have the manual and am looking through for ideas.

I am even thinking of pulling out the switches and cleaning each one individually as these have been culprits in other units I have had.



« Last Edit: March 08, 2023, 03:07:32 pm by Johnny10 »
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
 

Offline Gyro

  • Super Contributor
  • ***
  • Posts: 10773
  • Country: gb
Re: Tektronix 576 CRT Question
« Reply #1 on: March 06, 2023, 10:13:30 pm »
From what you say, I think it's unlikely to be a CRT problem (yet). Note that I am not a 500 series expert though.

If you have to turn the intensity way up to get a dot, it sounds as if the CRT grid bias is set wrong. It should be set so that, with the intensity control fully anticlockwise, the dot is just on the threshold of visibility (under dim room lighting) - that's on a scope anyway, it's presumably the same for a single dot on a curve tracer.

You say that you have checked all voltages. I suspect that these don't include all of the taps on the high voltage resistor chain, including the intensity, focus and astig controls. This is a high resistance chain that would be affected by the loading of the meter. The resistors in these chains are notorious for drifting, potentially taking the focus control out of correction range (and maybe causing the intensity problem too). It would be worth checking all of the values. I don't know if the thermal drift is fast enough to cause what you are seeing (again, it doesn't sound like a CRT problem though).

What you say about the Invert, zero and cal buttons introduces another possibility - noise pickup, but doesn't really tie in with the focus control observations.

Hopefully a couple of possibilities anyway.
« Last Edit: March 06, 2023, 10:15:04 pm by Gyro »
Best Regards, Chris
 

Offline Johnny10Topic starter

  • Frequent Contributor
  • **
  • Posts: 921
  • Country: us
Re: Tektronix 576 CRT Question
« Reply #2 on: March 06, 2023, 10:58:05 pm »
Thank you Gyro
I do know that there is a known problem with these switches that make the trace go wonky.
I am on my 5th Tek 576 but usually can get through the trace problems by spraying the 3 switches with a little Deoxit D5.
But so far no luck on this one.

Think like you said, I will start checking HV board resistors.

Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
 

Offline james_s

  • Super Contributor
  • ***
  • Posts: 21611
  • Country: us
Re: Tektronix 576 CRT Question
« Reply #3 on: March 06, 2023, 11:59:04 pm »
I've yet to see a CRT problem where it can be brought into focus and then drifts out. When CRTs get weak the most common symptom is a dim picture that gets blurry when the brightness is increased. This sounds more like a problem with the focus voltage or one of the grid voltages.
 

Offline jonpaul

  • Super Contributor
  • ***
  • Posts: 3859
  • Country: fr
  • Analog, magnetics, Power, HV, Audio, Cinema
    • IEEE Spectrum
Re: Tektronix 576 CRT Question
« Reply #4 on: March 07, 2023, 02:19:39 am »
high value resistance in CRT, focus, ckt, are known as drift up, over the decades.

Second suspect is HV transformers or multiplier.

Occasionally the CRT socket / neck to pins connections may be oxidized

Finally the 576 CRT same,as  7603 and other Tektronix large display CRTs.

Bon Soirée

Jon

The Internet Dinosaur..
passionate about analog electronics since 1950s
 
The following users thanked this post: james_s

Offline Johnny10Topic starter

  • Frequent Contributor
  • **
  • Posts: 921
  • Country: us
Re: Tektronix 576 CRT Question
« Reply #5 on: March 08, 2023, 03:03:50 pm »
Looks like the problem is Intensity and Focus pots.

I compared ohms reading with my working unit and the readings are way off.
Looking for replacements now.

var. potentiometers are Tektronix part # 311-0254-00 and 311-0397-01
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
 

Offline james_s

  • Super Contributor
  • ***
  • Posts: 21611
  • Country: us
Re: Tektronix 576 CRT Question
« Reply #6 on: March 08, 2023, 05:40:45 pm »
Are they constructed such that you can clean them out with some control cleaner?
 

Offline Johnny10Topic starter

  • Frequent Contributor
  • **
  • Posts: 921
  • Country: us
Re: Tektronix 576 CRT Question
« Reply #7 on: March 08, 2023, 08:06:38 pm »
I haven't pulled them out yet.
Measured the 2M ohm pot with wires removed, I believe the pots are sealed, these are H.V. units with insulated shafts.

I found new replacements, after removing wires and turning the 2M Ohm pot through its range, it read from 15 Ohm - 38M Ohm - 2.54 M Ohm.
The good pot read from 1.0 Ohm smoothly to 1.82 M Ohm.

Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
 

Offline james_s

  • Super Contributor
  • ***
  • Posts: 21611
  • Country: us
Re: Tektronix 576 CRT Question
« Reply #8 on: March 08, 2023, 08:13:59 pm »
It might just be dirty. Try squirting some control cleaner down the shaft, enough might get inside. Turn the pot back and forth a couple dozen times and then measure it again.
 

Offline Johnny10Topic starter

  • Frequent Contributor
  • **
  • Posts: 921
  • Country: us
Re: Tektronix 576 CRT Question
« Reply #9 on: March 08, 2023, 09:40:23 pm »
Well, I took the control out and took it apart. It was not sealed and it had cleaner in the pot.
After drying it out and putting it back together it was the same jumped from 15 Ohm all the way to 38M Ohm and higher before reaching out of range at the end.
The track inside was completely smooth.

I think it best to replace.
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
 

Offline james_s

  • Super Contributor
  • ***
  • Posts: 21611
  • Country: us
Re: Tektronix 576 CRT Question
« Reply #10 on: March 08, 2023, 10:07:23 pm »
That's weird, pots in that sort of application don't usually wear out.
 
The following users thanked this post: Johnny10

Offline Johnny10Topic starter

  • Frequent Contributor
  • **
  • Posts: 921
  • Country: us
Re: Tektronix 576 CRT Question
« Reply #11 on: March 09, 2023, 12:38:59 pm »
Someone tried to repair this unit before my purchase. ie, screws missing, high voltage cover missing.

The switch I took apart was filled with liquid, I don't know what it was. Although it was extremely slippery and would not dry out.

The main part of track was completely smooth with grainy track only on two ends.

I may take the other one apart also and see.
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
 

Offline Johnny10Topic starter

  • Frequent Contributor
  • **
  • Posts: 921
  • Country: us
Re: Tektronix 576 CRT Question
« Reply #12 on: March 09, 2023, 05:23:29 pm »
OK James,
I thought about what you said, So I took both switches back apart and soaked them in alcohol to remove whatever residue was on the tracks and now I have a focused trace again. The numbers are not perfect but it definitely made for a temporary workable solution.

I will still replace those switches.
Now to find any more problems.

I set up my work bench again and in the past few days have started fixing some of my equipment.

Thank for your help. I haven't been doing much repair for a couple of years now so it is always nice to present the problem.
« Last Edit: March 11, 2023, 12:38:42 pm by Johnny10 »
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
 

Offline james_s

  • Super Contributor
  • ***
  • Posts: 21611
  • Country: us
Re: Tektronix 576 CRT Question
« Reply #13 on: March 09, 2023, 06:24:23 pm »
Sounds like maybe somebody used the wrong type of product in attempt to clean them.

Does the image look good now with the pots cleaned?
 

Offline Johnny10Topic starter

  • Frequent Contributor
  • **
  • Posts: 921
  • Country: us
Re: Tektronix 576 CRT Question
« Reply #14 on: March 11, 2023, 12:39:15 pm »
Posting pic, after cleaning switches.
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
 

Offline jonpaul

  • Super Contributor
  • ***
  • Posts: 3859
  • Country: fr
  • Analog, magnetics, Power, HV, Audio, Cinema
    • IEEE Spectrum
Re: Tektronix 576 CRT Question
« Reply #15 on: March 11, 2023, 01:11:29 pm »
very good, but seems like the intensity is rather high

we minimize the intensity as,needed.

Burning is,easy on curve tracer!

Jon
The Internet Dinosaur..
passionate about analog electronics since 1950s
 

Offline Johnny10Topic starter

  • Frequent Contributor
  • **
  • Posts: 921
  • Country: us
Re: Tektronix 576 CRT Question
« Reply #16 on: March 11, 2023, 03:57:38 pm »
Yes, I ordered new switches, trace drops out while bringing intensity down.
The resistance of these switches is not linear. Need to be replaced.
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
 

Offline james_s

  • Super Contributor
  • ***
  • Posts: 21611
  • Country: us
Re: Tektronix 576 CRT Question
« Reply #17 on: March 11, 2023, 10:00:10 pm »
It's remarkable that there does not appear to be any burn on that tube already, most of these curve tracers I've seen have a dark spot in the corner.
 

Offline Johnny10Topic starter

  • Frequent Contributor
  • **
  • Posts: 921
  • Country: us
Re: Tektronix 576 CRT Question
« Reply #18 on: March 11, 2023, 10:36:24 pm »
Yes and it is a very early unit.
All IC's are 1968 or before.

Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
 

Offline jonpaul

  • Super Contributor
  • ***
  • Posts: 3859
  • Country: fr
  • Analog, magnetics, Power, HV, Audio, Cinema
    • IEEE Spectrum
Re: Tektronix 576 CRT Question
« Reply #19 on: March 12, 2023, 03:18:06 am »
Johnny10, congrats, 576, is a,fine,machine

I.suspect that theliquid filled  pot was,lubed with WD40,by a,novice.

Check,all other posts and trimmers,for similar abuse...

Fail points on 576 are,HV transformers, mult, switches and fiber optic readout lamps.

Look for the rare front porches on,ebay or,ham radio Fleas être are several besides the stock

100A pulsed
programmable
SCR current

Tektronix numbers like
172, 176

The service manual is also  very useful.

 Jon

The Internet Dinosaur..
passionate about analog electronics since 1950s
 
The following users thanked this post: Johnny10

Offline Johnny10Topic starter

  • Frequent Contributor
  • **
  • Posts: 921
  • Country: us
Re: Tektronix 576 CRT Question
« Reply #20 on: March 12, 2023, 12:20:55 pm »
That's a good thought jonpaul.

If wrong cleaner or lube in those switches possibly others also.
Tektronix TDS7104, DMM4050, HP 3561A, HP 35665, Tek 2465A, HP8903B, DSA602A, Tek 7854, 7834, HP3457A, Tek 575, 576, 577 Curve Tracers, Datron 4000, Datron 4000A, DOS4EVER uTracer, HP5335A, EIP534B 20GHz Frequency Counter, TrueTime Rubidium, Sencore LC102, Tek TG506, TG501, SG503, HP 8568B
 

Offline jonpaul

  • Super Contributor
  • ***
  • Posts: 3859
  • Country: fr
  • Analog, magnetics, Power, HV, Audio, Cinema
    • IEEE Spectrum
Re: Tektronix 576 CRT Question
« Reply #21 on: March 12, 2023, 02:12:17 pm »
Rebonjour Cher monsieur Johnnie

As power electronics designer, consultant and manufacturing in 1970s..1990s, we used both 577 and 576 Tektronix extensively.  Tested 20,000 150V HV Zener diodes for production avionics power supply.

Very lucky to find these curve tracers in 1980s, in excellent condition with test fixtures and front porch 100A pulsed, programmable and regular. At Silicon Valley ham radio Fleas and auctions.

1. use 99% or 91 % isopropyl alcohol to clean test equipment knobs, panel's, PCBs, etc.  Very low risk of damage

2.please  DO NOT use WD40, petroleum distillates, or powerful organic solvent like acetone, trichlorethene, etc severe  damage to plastic, components etc.

3. Many forum vintage equipment folk are afficanados of Craig Deoxit products for controls and pots. 

4. We have managed for decades to use ordinary TV tuner cleaner, control lube etc. from Miller Stephenson, GC, etc. Some tektronix equipment have sealed square blue Bourne pots, more difficult to clean.

5. Additional common failures of 576 to beware of are the deteriorated plastic on certain knobs and controls, e.g.  the small top row focus, intensity controls.

Also, especially delicate  are the main large control knobs for mode, R/ power/Voltage range, etc.

Very   easy to split knob at sets screw and nearly impossible to repair or to find replacement knobs. ROTATE LARGE CONTROLS SLOWLY, and  CAREFULLY!

6. Besides eevblog, best advice, parts, accessories on all Tektronix scopes and curve tracer at tekscopes2 and tekscopes @ groups.io, some veteran Tektronix engineering and technicians...Best information, manuals at W140 and Tekwiki.

7. Test fixtures are rare and costly, but DIY versions and PCBs are on both eevblog and tekscopes2
Our most used  fixtures are TO3, TO247, TO220, diode 4 terminal, and TO92.

8. The high voltage  safety cover is useless and easy to break. Most simply defeat the interlock switch at lower left of front porch.

Hope you enjoy my notes and tips...

Just the ramblings of an old retired EE (1968!)

Jon

The Internet Dinosaur..
passionate about analog electronics since 1950s
 