Author Topic: Tektronix CSA-8000 / TDS-8000 repair experiences?  (Read 19568 times)
Offline Jwalling
  • This is work?
Re: Tektronix CSA-8000 / TDS-8000 repair experiences?
« Reply #25 on: June 26, 2017, 03:58:11 pm »
http://www.ebay.com/itm/TDS6000-TDS7000-TDS8000-Windows-7-Upgrade-Kit-with-chassis-TDS6000-TDS7000-TDS-/401335497819?hash=item5d7175b45b:g:BSkAAOSwTf9ZThVn Look at that? It seems it works? But I don't know why you would need Win 7 because from what I thought all of the newest updates are for XP.
Simon
That's nuts.  :palm:
https://support.microsoft.com/en-ph/help/10737/windows-7-system-requirements
MS recommends 1GB RAM minimum and 1GHz CPU.
When you start loading analysis software (And I've seen some that will use over 300MB of memory) you're going to use up all the memory and be using the swap file on the disk drive. Performance must suck!
Jay
System error. Strike any user to continue.
Offline tpsimon
« Reply #26 on: June 26, 2017, 07:10:03 pm »
This makes no sense to me, but I installed 128Mb of ECC SDRAM into the motherboard (That's it!), ran SP2, and right now XP is installed. I will run more thorough tests with the OS, but it's booting, and seems to be running like a champ on my current setup. :box: I will mention though, I have another partition on the HDD for XP, and have the old '98 setup on another partition as well. (This keeps the PowerPC board happy). I can't seem to figure out how to install the Tek 8000 Series Oscilloscope Application, as I got TekVisa Installed and all of that. I guess I'll go fish for it on the disk.
I'll keep you posted,
Simon
Offline EB5AGVTopic starter
  • The best recycling is repairing
    • Amateur Radio and Test Equipment
« Reply #27 on: June 26, 2017, 09:46:12 pm »
Maybe this one is the one you were looking for:
http://www.ebay.com/itm/Tektronix-TDS7054-TDS7104-TDS-CSA7404-TDS-CSA8000-Windows-XP-Upgrade-Opt-Chassis-/182637750824?epid=96961873&hash=item2a860eea28:g:tyYAAOSwMmBV1pBR
Simon
Yes, but that one is for XP. In fact, the kit I have got is one of these, but I didn't pay for it directly. It is a long story...
Jose
JOSE V. GAVILA - La Canyada - Valencia (SPAIN)
http://agvradio.com  AGVradio
http://jvgavila.com   Personal WEB
Offline tpsimon
« Reply #28 on: June 26, 2017, 10:55:56 pm »
So I installed the drivers for the 7k series on my scope and edited my PowerPC parameters to match the new firmware, but the scope won't load the scope program. Any ideas? (I successfully installed the v5.2 Software)
Thanks!
Simon
« Reply #29 on: June 26, 2017, 10:58:17 pm »
Thanks!
Simon
How does the POWER PC mainboard log compares to the one I published on this thread?. As my unit boots and works on XP, that could be a good lead.
Jose
Offline tpsimon
« Reply #30 on: June 26, 2017, 11:19:08 pm »
Hey,
Mine looks very similar. (I am just using D drive instead of C)
UUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUUU
romPost: MPC106  Walking-one passed.
romPost: MPC106 configuration passed.
romPost: First PCI access  passed.
romPost: PC87560 Walking-one passed.
romPost: PC87560 configuration passed.
romPost: DEC21554 configuration passed.
romPost: PERFORMING ROM CHECKSUM TEST...
romPost: ROM checksum passed.
romPost: PERFORMING DRAM CELL TEST...
romPost: Enabling data cache...
romPost: Data cache enabled.
romPost: Enabling instruction cache...
romPost: Instruction cache enabled.
romPost: DRAM Cell test passed.
romPost: Disabling data Cache...
romPost: Data cache disabled.
romPost: Disabling instruction cache...
romPost: Instruction cache disabled.
romPost: PERFORMING DRAM MARCH TEST WITH CACHE OFF...
romPost: Writing address to each *address...
romPost: Reading address from each *address...
romPost: DRAM march test passed...
romPost: PERFORMING DRAM MARCH TEST WITH CACHE ON...
romPost: PERFORMING DRAM WALKING-ONE  TEST...
romPost: DRAM Walking-one test passed.
romPost: PERFORMING NV-MEMORY WALKING-ONE  TEST...
romPost: NV-MEMORY Walking-one test passed.
romPost: POWER-UP DIAGNOSTIC COMPLETE.
Hel
                            VxWorks System Boot
Copyright 1984-1996  Wind River Systems, Inc.
CPU: Tektronix HPP - PowerPC 740
Version: 5.3.1
BSP version: 1.0.3.0
Creation date: Jun 28 2000, 16:52:12
Press any key to stop auto-boot...
 1
 0
auto-booting...
boot device          : sm
processor number     : 0
host name            : host
file name            : d:\progra~1\tdscsa~1\firmware\vxWorks.st
inet on ethernet (e) : 192.168.0.2:ffffff00
inet on backplane (b): 192.168.0.2:ffffff00
host inet (h)        : 192.168.0.1
gateway inet (g)     : 192.168.0.2
user (u)             : anonymous
ftp password (pw)    : anonymous
flags (f)            : 0x1008
target name (tn)     : target
startup script (s)   : d:\progra~1\tdscsa~1\firmware\dio.vsh
other (o)            : nvfs=0x1000
Initializing backplane net with anchor at 0x1e80000... done.
Backplane anchor at 0x1e80000... Attaching network interface sm0... done
done.
Attaching network interface lo0... done.
PC1 dio device driver installed
Attaching DIO client... done.
Loading d:\progra~1\tdscsa~1\firmware\vxWorks.st...1771964
Starting at 0x200000...
Initializing backplane net with anchor at 0x1de0000... done.
Backplane anchor at 0x1de0000... Attaching network interface sm0... done
done.
Adding 6006 symbols for standalone.
      ]]]]]]]]]]]  ]]]]     ]]]]]]]]]]       ]]              ]]]]         (R)
 ]]]]]]]]]]]]]]]]]]]]]]]]]]]]]       Development System
 ]]]]]]]]]]]]]]]]]]]]]]]]]]]       VxWorks version 5.3.1
 ]]]]]]]]]]]]]]]]]]]]]]]]]]       KERNEL: WIND version 2.5
 ]]]]]]]]]]]]]]]]]]]]]]]]]       Copyright Wind River Systems, Inc., 1984-1997
                               CPU: Tektronix HPP - PowerPC 740.  Processor #0.
                              Memory Size: 0x2000000.  BSP version 1.0.3.5.
                             WDB: Ready.
usrNvFsInit: NAME=/nvram BASE=0xfd0ff100 BPB=512 BPT=8 NB=8 BO=0
nvfs: STATUS: Mounting existing NVRAM file system
KERNEL: THERMAL CONTROL INITIATED
Executing startup script d:\progra~1\tdscsa~1\firmware\dio.vsh ...
dbFsIoctl: invalid code 3
cd "c:/progra~1/tdscsa~1/firmware"
value = 0 = 0x0
*0xfec00000 = 0x04b00080;*0xfee00000 = 0x17009002
ppp_parms + 0xfe84f680 = 0xfec00000: value = 78643328 = 0x4b00080
ppp_parms + 0xfea4f680 = 0xfee00000: value = 385912834 = 0x17009002
*0xfec00000 = 0x88a10080;*0xfee00000 = 0x01800000
ppp_parms + 0xfe84f680 = 0xfec00000: value = -2002714496 = 0x88a10080 = ppp_parms + 0x8865f700
ppp_parms + 0xfea4f680 = 0xfee00000: value = 25165824 = 0x1800000
*0xfe008002 = 0;*0xfe00800e = 0;*0xfe008012 = 0
ppp_parms + 0xfdc57682 = 0xfe008002: value = 0 = 0x0
ppp_parms + 0xfdc5768e = 0xfe00800e: value = 0 = 0x0
ppp_parms + 0xfdc57692 = 0xfe008012: value = 0 = 0x0
symSyncTimeoutSet 600000
value = 600000 = 0x927c0
ld <hi32.o
value = 31002016 = 0x1d90da0 = Hi32SystemProcessorBoard + 0xa98
Hi32SystemProcessorBoard = 1
Hi32SystemProcessorBoard = 0x1d90308: value = 1 = 0x1
CmdDspDevName = "/dev/cmddsp0"
new symbol "CmdDspDevName" added to symbol table.
CmdDspDevName = 0x1d90cc0: value = 31002464 = 0x1d90f60 = CmdDspDevName + 0x2a0
hi32Drv
value = 0 = 0x0
hi32DevCreate CmdDspDevName,0x80006000,0xa0000000,3
value = 0 = 0x0
ld <cmddsp.o
value = 31000320 = 0x1d90700 = Hi32SystemProcessorBoard + 0x3f8
ld <acqdsp.o
value = 31000800 = 0x1d908e0 = Hi32SystemProcessorBoard + 0x5d8
ld <diagcmddsp.o
value = 31008192 = 0x1d925c0
ld <diagModule_wmp.o
value = 31007552 = 0x1d92340
ld < systemLibraryCode.o
value = 31007104 = 0x1d92180
CSSDefaultDioDisable=0
CSSDefaultDioDisable = 0x1b20480: value = 0 = 0x0
semInitForQueueCheckLogMsg
value = 31001312 = 0x1d90ae0 = Hi32SystemProcessorBoard + 0x7d8
ld < nvSysLib.o
value = 30999648 = 0x1d90460 = Hi32SystemProcessorBoard + 0x158
ld < bucket.o
value = 31003968 = 0x1d91540 = CmdDspDevName + 0x880
MathMeasCreateMappingFcns = 1
MathMeasCreateMappingFcns = 0x1aa60b8: value = 1 = 0x1
TmgrExecuteAllMappingFunctionsEnable=1
TmgrExecuteAllMappingFunctionsEnable = 0x1aa2694: value = 1 = 0x1
ld < as.o
value = 28451456 = 0x1b22280
ld < ifcProxy.o
value = 28450880 = 0x1b22040
acqCheckForCmdDsp=1
new symbol "acqCheckForCmdDsp" added to symbol table.
acqCheckForCmdDsp = 0x1d90660: value = 1 = 0x1
ld < rtlLibTarget.o
value = 28340096 = 0x1b06f80
ld < slotMgrLightLibTarget.o
value = 28287200 = 0x1afa0e0
sysExtMemAddToPool memSysPartId ,0x02000000 ,0x02000000
value = 0 = 0x0
MathMeasDisableParseExpr = 1
new symbol "MathMeasDisableParseExpr" added to symbol table.
MathMeasDisableParseExpr = 0x3fff400: value = 1 = 0x1
use100Breakpoints=1
use100Breakpoints = 0x14f6e34: value = 1 = 0x1
pWfmMemPool=0
pWfmMemPool = 0x1abf1f0: value = 0 = 0x0
AcqLibFourChannelProto = 0
new symbol "AcqLibFourChannelProto" added to symbol table.
AcqLibFourChannelProto = 0x3fff3a0: value = 0 = 0x0
AcqExecHardWireChannelSources = 0
AcqExecHardWireChannelSources = 0x119ecfc: value = 0 = 0x0
doTekprobeExec=1
doTekprobeExec = 0x128cbc4: value = 1 = 0x1
pDbDev = ramDevCreate(0, 640, 1024, 1024, 0)
new symbol "pDbDev" added to symbol table.
pDbDev = 0x3f5f2c0: value = 67105536 = 0x3fff300
pDBfs  = dosFsMkfs("/db", pDbDev)
new symbol "pDBfs" added to symbol table.
pDBfs = 0x3f5dba0: value = 66449888 = 0x3f5f1e0
copy ("TDS666.DBF", "/db/TDS666.DBF")
value = 0 = 0x0
dbPath = "/db/"
new symbol "dbPath" added to symbol table.
dbPath = 0x3f5daa0: value = 66444064 = 0x3f5db20 = dbPath + 0x80
wfmMemInit 0
value = 55695744 = 0x351d980
jNVInitMgrs
value = 0 = 0x0
TmgrInit(dbPath)
value = 0 = 0x0
rtlInit
value = 0 = 0x0
MathMeasCreateTheMathandMeasurementLibraries
value = 50112544 = 0x2fca820
initAcqMapper
value = 50083584 = 0x2fc3700
autosetMapInit
value = 50083168 = 0x2fc3560
mathMapperInit
value = 0 = 0x0
initUtilityMapper
value = 50076320 = 0x2fc1aa0
initWfmXmtrMapper
value = 50072160 = 0x2fc0a60
commInit__Fv
value = 0 = 0x0
JanusProxyInit
value = 50038272 = 0x2fb8600
acqExecAllocUploadBuffers
undefined symbol: acqExecAllocUploadBuffers
Done executing startup script d:\progra~1\tdscsa~1\firmware\dio.vsh
KERNEL:  RUNNING TIME SLICED AT 10 TICKS PER SLICE!
But I think I see what's wrong. I'll copy the OS over to a fresh drive and let you know, because everything is on the D: drive right now, and it seems that the code needed it to be on the C: drive.
Simon
Edit:
It seems to make no difference if I change the .vsh file.
Edit 2:
I will just wait for the board to come in tomorrow and see what it does.
« Last Edit: June 27, 2017, 01:41:09 am by tpsimon »
Offline tpsimon
« Reply #31 on: June 28, 2017, 01:18:52 am »
Got the whole scope working! Getting some Optical Modulators tomorrow so that I can mess with the scope.
Simon
Offline ID10TError
« Reply #32 on: June 28, 2017, 01:50:16 am »
Hello Everyone,
My boss bought a CSA8000 off of eBay that had a failed hard drive. He asked me to put a new SATA hard drive in it and install Windows 2000. I got a little back plane converter to put in a SATA drive in the place if the IDE. Here is what I have so far. The unit came to me, failed hard drive, no video on the display (works with external monitor), drive B error at boot up, PXE boot (no hard drive seen), says it will enter setup when F2 is pressed (never goes into setup). CD player does not open.
So, I downloaded a BIOS update for Intel BL440ZX board from the Intel site. Put it on a bootable floppy and booted the system up with the floppy. It came up and ran the BIOS update (big mistake). Now when I try to boot it up, there is nothing. No display, no boot sounds, no post, etc. I remove the CMOS battery, nothing. I set the CMOS jumper to pins 2/3 from 1/2, still nothing.
I read somewhere after the fact not to update the BIOS using firmware updates from the Intel site. That is what I did. How can I get this thing back with the right BIOS firmware so it will somewhat boot again?
PS - If I take the jumper off the pins and power on, it will try to read the floppy drive. Won't boot from it, but I believe that setting is to restore a BIOS if I had the proper files.
Thanks for your help!
ID10T Error
« Reply #33 on: June 28, 2017, 07:49:31 am »
Simon
Great news!!!
Please, try the compensation procedure for the mainframe (and any plug-in you may have) and let us know the results. If the unit passes, you have got a good one  :-+
Regards,
Jose
Offline ID10TError
« Reply #34 on: June 28, 2017, 08:01:38 pm »
Anyone have any thoughts on how I can re-flash the BIOS and where I can get the required BIOS files?
Thanks,
« Reply #35 on: June 28, 2017, 08:30:39 pm »
Thanks,
Sorry, I have not (yet!) needed to do that on any of my units. All other possible faults have happened, though  |O
One possible solution is to get a replacement mainboard. It would be cheap enough, I guess, as when you update to Windows XP there is a surplus mainboard left. I have one but I live in Spain, a bit far from you.
Regards,
Jose
« Last Edit: June 29, 2017, 08:12:38 am by EB5AGV »
Offline ID10TError
« Reply #36 on: June 28, 2017, 09:20:42 pm »
Thanks Jose!
Yeah, I live in the States. A bit far to ship. I wished I wouldn't of tried updating the BIOS. It was posting, although I couldn't go into setup or see the drives. I contacted Tektronix and they don't seem to be able to help me. All they say is to ship the unit in for repair. Not sure what that would cost. I want to put Windows 2000 on it. I tried everything I could think of. I don't get any display now. It would be nice if we could back out the BIOS update I did. It had to be the right update for that board or I wouldn't think it would have taken it.
Would anybody be able to take a backup of their BIOS and send me the files? It would be greatly appreciated!
Thanks!
Offline tpsimon
« Reply #37 on: June 29, 2017, 01:47:49 am »
@EB5AGV Compensation passes, everything passes! I guess I lucked out!!! I ran all the tests. The only thing is that one of these modules that I bought hasn't been calibrated since 1999! I'm sure there's a calibration procedure somewhere. (I heard it's really well documented)
@ID10TError Did you hook up an external display (The top port)? If there is any output I can guide you through that because I had to go through that procedure, it took many many hours to figure out. (I also saw on another forum DO NOT FLASH INTEL BIOS, so I didn't, but hey maybe it will still work because my XP machine runs the Intel BIOS)
Offline ID10TError
« Reply #38 on: June 29, 2017, 01:47:40 pm »
@tpsimon - I do have an external monitor hooked up because the built in display does not appear to be working. However, I get no display, post, beeps, nothing. If I remove the CMOS jumper from all the pins, when I turn it on, you can hear it try to access the floppy drive, but still no display. I read somewhere where it would read a disk with the correct backup BIOS on it and reload it. I was hoping to find someone with a good BIOS they could make a backup copy for me and I can see if that will load. Any thoughts?
Offline tpsimon
« Reply #39 on: June 29, 2017, 05:08:13 pm »
@ID10TError Unfortunately I do not have a copy of the software BIOS for that Motherboard Model. My unit did not come with any backup CDs etc. You can grab a new motherboard with all the CDs on Ebay; that is probably your best bet.
Simon
Offline casinada
« Reply #40 on: June 29, 2017, 08:39:14 pm »
go to Ilya's web site: (Tin)  :-+
https://xdevs.com/fix/csa7404/
and read carefully. You might find the information you're looking for.  :)
Offline frederic73
« Reply #41 on: July 06, 2017, 03:15:33 am »
Hi everyone
I have been given the Task to repair a TDS 8000. I'm an IT person btw.
So far i have replaced the cpu, upgrade to 512Mb and installed windows Xp . it seems the MB is the SU 810. Most drivers had been installed so far
I have read all the pages i could find on the web and this forum. The Tek scope application will hangs when launched , so i am getting a new battery for the power pc.
What is my next step to make sure the nvram is functional and power pc boots again. I am reading this page https://www.eevblog.com/forum/testgear/tek-csa7404-repair-project/
but is confuse about how to get a serial connected and also  load the proper powerpc firmware/booltlaoder
Thanks for your help. :)
Offline alittleontheheaviside
« Reply #42 on: August 29, 2017, 07:32:43 am »
Hi guys,
I've got a sick puppy (the harddisk is flapping about) of a CSA8200 (same family as TDS8000 series).
Can anyone point me in the right direction for the reinstall windows.... doesn't appear available from the tek site.
much appreciated!
Offline PrecisionAnalytic
« Reply #43 on: December 27, 2017, 08:39:38 pm »
Hi guys,
much appreciated!
This is an older post... though for the record  :popcorn::
You can place the HD in a zip lock or sealed freezer bag and freeze then take out and backup and/or clone the drive to save time on re-installation.   You can repeat this until you recover the data.  Sure beats hunting down an exact board for the drive and replacing to recover the data... that is... if the board is bad.
If you want, you can perform error checking also.
I'm prepping to do this before I open the case of the TDS-8000B I purchased for a great deal and am hoping to restore to use as the "oscilloscope I don't want to use too much, since costs so much, though is my best scope."   :D
Offline GastonBA
« Reply #44 on: August 25, 2018, 09:29:38 pm »
hello,
could you help? I am having drivers issues trying to reconstruct from winXP. Could you help?
Offline GastonBA
« Reply #45 on: September 02, 2018, 04:52:51 am »
Hey, could you guide into what you did to have XP working?
I am suspecting that the PCI1225 is part of the problem that has conflicts on my unit (no enough resources).
Thank you,
Gaston
Offline GastonBA
« Reply #46 on: September 02, 2018, 04:54:59 am »
Jose, could you help? I am stuck on XP upgrade.
Gaston
Offline Jwalling
  • This is work?
« Reply #47 on: September 02, 2018, 09:15:43 am »
Thank you,
Gaston
I think the PCI1225 MAY be the PCMCIA controller. If so, that's not a problem. Check the vendor ID in the device manager - if it's TI then don't worry about it, your problem lies elsewhere.
Jay
Offline rizal5478
  • Country: my
« Reply #48 on: April 20, 2020, 01:20:15 pm »
Hi Jay, Recently happen to my unit CSA8000 which I can't enter to boot setup and keep on error show drive A & B drive error. For 1st time shall I run your corrupt.zip through floppy disk? Thanks in advance for your advise.
Offline Jwalling
  • This is work?
« Reply #49 on: April 21, 2020, 09:27:02 am »
Yes, that will allow you to get into the BIOS setup again. One of the most idiotic BIOS bugs I've ever seen...
You will need to connect a VGA monitor to the NLX motherboards VGA port, then set the default video from AGP to PCI in the BIOS.
Jay