Tektronix Current probe DC Drift
« on: August 25, 2021, 04:58:58 am »
Hello I came across a TCP00030 Current probe.  The probe  has a DC drift of about 70mV, until it warns up. Once warned up it settles within several mV of  the bassline of zero. Is the drift cause by a defective hall effect sensor,  down stream electronic or a mechanical part in the handle? :-//
Any suggestions  where to start.

thanks
 

Offline Hamelec

  • Regular Contributor
  • *
  • Posts: 154
  • Country: de
Re: Tektronix Current probe DC Drift
« Reply #1 on: August 25, 2021, 08:02:38 am »
nothing at all.
But see instruction manual:
"Quick Tip:
To maintain measurement accuracy, degauss your probe in each of these cases:
After you turn on the measurement system and allow a 20-minute warm-up period"
« Last Edit: August 25, 2021, 08:09:18 am by Hamelec »
 

Offline gogomanTopic starter

  • Regular Contributor
  • *
  • Posts: 116
Re: Tektronix Current probe DC Drift
« Reply #2 on: August 26, 2021, 12:37:44 am »
An identical TCP drift 1-2mA at power up after which it locks on to the baseline settings. Which leads me to believe the drifty one has a problem .

thanks
 