Author Topic: TEKTRONIX DM5010 and FG5010 Switches Sticking  (Read 891 times)
Offline cjuriedTopic starter
    • Juried Engineering
« on: October 14, 2023, 02:14:56 pm »
Hi all,
I have a DM5010 and FG5010 which have momentary pushbuttons that are sticking. Is there a recommended cleaning procedure to resolve this issue? Is this a known  common problem with these modules?
Thank you.
juriedengineering.com/
Online coromonadalix
Re: TEKTRONIX DM5010 and FG5010 Switches Sticking
« Reply #1 on: October 14, 2023, 04:49:26 pm »
not common as the racal dana ones,  but over time,  since they are old systems,
the switch components inside may degrade, i have an dm5120  who began to exhibit the same symptoms
and finding substiutes .... some have leds inside them in my case
you have to measure their heights and sizes ... and try to find similars ones
user manual and SV manual here
https://w140.com/tekwiki/wiki/DM5010
https://w140.com/tekwiki/wiki/FG5010
« Last Edit: October 14, 2023, 04:56:27 pm by coromonadalix »
« Reply #2 on: October 14, 2023, 05:57:43 pm »
Indeed. However is there a cleaning solvent and/or lubricant procedure that has been shown to resolve this issue?
« Reply #3 on: October 15, 2023, 01:55:58 am »
deoxit can do wonder for oxidized contacts
but if the mechanical spring action is foam based or plastic dome / metal dome ....  degradation / deformation over time can happen
lubricant : i would not try to use them,    you have to dry clean all contacts to be sure, deoxit them, let them dry, or use compressed air to move everything out and do some tests
petroleum based lubricant could do more damaged than good, you have to use dedicated ones for switchs ??
« Last Edit: October 15, 2023, 01:58:39 am by coromonadalix »
Online TERRA Operative
  • Country: jp
  • Voider of warranties
    • Near Far Media Youtube
« Reply #4 on: October 16, 2023, 12:07:27 pm »
In my AFG5101 IIRC I carefully removed the cap and cleaned the sticky ones with isopropyl alcohol them clicked the top back on.
Was a while ago though, so I can't remember the details beyond that unfortunately.
« Reply #5 on: October 17, 2023, 12:14:30 pm »
Was it a disassembly of the front panel from the main chasis or were you able to remove the caps with the front panel attached? Thank you.
  • Country: jp
« Reply #6 on: October 17, 2023, 01:37:12 pm »
It wasn't too hard to remove the front panel. A few screws on the top and bottom and the whole front assembly came away, then unplug a few cables and then remove the screws holding the PCB in place and it came apart.
« Reply #7 on: October 17, 2023, 02:54:59 pm »
I will give it a try. Thank you.