| Electronics > Repair |
| Tektronix DM501A repair |
| << < (5/7) > >> |
| Kleinstein:
The multiplexter in the LD120 would be more a thing of thermal drift and offset, not occasional wandering. That is more a contract / switch isssue. For the MUX replacement one does not need low R_on, but more low leakage. There is a large resistor in series anyway. So the DG419 is more on the low R-on side, but an lower leakage alternative like DG213 usually comes with more switches and pins. |
| David Hess:
--- Quote from: Rax on May 30, 2024, 10:34:40 pm ---Any reason for choosing an "11 ohm through resistance" part? By which I mean the DG419. The 417 has 6 ohms, I'd be inclined to think less resistance is better here?... --- End quote --- The DG419 was the lowest leakage part that I could easily get at the time. The Texas Instruments TMUX6119 should be better. The DIP package DG419 does make it easier to air-wire however for lowest leakage. |
| Rax:
--- Quote from: Kleinstein on May 31, 2024, 09:22:11 am ---The multiplexter in the LD120 would be more a thing of thermal drift and offset, not occasional wandering. That is more a contract / switch isssue. --- End quote --- I agree. I'll persist on conditioning the switches - hopefully, they're feasibly fixable. |
| Rax:
--- Quote from: David Hess on May 06, 2024, 12:54:08 am ---I have had excellent results with MG Chemicals 801B - Super Contact Cleaner With PPE (Polyphenyl Ether). Its solvent is naphtha which is plastic safe, and the PPE oil does not creep across surfaces unlike silicone oils minimizing contamination. --- End quote --- From my experience with it this far with 801B on these switches, it really needs a generous amount of time to work and cure (have the volatile parts evaporate, etc.). Differently put, it seems to me it took a day or maybe a couple before I saw the meters behave as expected (nominally). Before that I saw all sorts of leakage and slightly weird readings. Not sure what your experience from this perspective is, and my own data is still trickling in. |
| Rax:
On further thoughts with these units, I came to find what everyone has been saying - the switches on these meters are a VERY weak point. And I don't capitalize easily. I can't really fathom why Tektronix used such cheap, flaky switches. I've not seen similarly underwhelming parts with competitors - HP, etc. This far, I have one unit out of five I regard now as fully functional - a couple others waiting for 801B to do its work - but given the time they need to have their switches fixed it's by a long shot a labor of love. |
| Navigation |
| Message Index |
| Next page |
| Previous page |