Hi guys,
Currently I'm resurrecting a high-end scope (especially back in the days) TEKTRONIX DPO 70604C. After replacing its HDD and ATX PC board it finally boots (it fails to do so from time to time, but whatever...).
Right mow, however, I'm facing acquisition board issue. On the first channel, after switching the full sampling rate (from 12,5 GS/s to 25 GS/s) I see large spikes appearing. This tells me, that some additional sampling circuit kicks in and starts to show bad behavior. There is 320 ps of separation between the spikes, thus it is 25 GS/s divided by 8. Makes sense - there are multiple ADC for a single channel for sure.
I tracked it a bit more and found a chip under glued heatsink - see arrow in the picture. This is probably the CH1 multiplexer. After pushing it, especially on one of the crones as my pencil shows, or by spraying some freezer - the spikes stop and scope behaves as new. Is this a broken ball behavior?
If it is, is it worth to try heating up and pray for the IC to reach the PCB? I am quite afraid about doing so, because:
1) The board is about 32 layers and 5 mm of thickness.
2) The thermal addhesive underneath the heatsink is solid as a rock - very hard and quite impossible to remove. I tried tu cut it a bit with a dental floss but without much luck.
3) The suspect IC is surrounded by other delicate ADC ICs.
Is it worth trying to re-solder this IC?
Best!
Mateusz