Replaced the TOP244GN chip and put everything back together and the scope works like a charm again

here are some information for those who might need it in the future:
as soon as you plug in the scope, before turning it on, a Green LED lights up near the CMOS battery (above the ADCs)
on J0401 connector (7 pins coming from the PSU) we have these voltages:
pin 3 is 3.5000V and pin 4 is 5.18V.
after powering the scope on, the 3.5000V disappears but pin 6 now has 3.22V. Pin 4 still has 5.17V
also after power on 2 more green LEDs light up on the motherboard
the main voltages on the big PSU connector are 6.03V (red) and 12.6V (orange)
on the PSU board itself, the TIP244 chip is ON right after plugging the scope into outlet. It generates 3 voltages.
1) 13.16V on C90 which supplies the LM5033 (the main DC-DC converter) and other stuff on the primary side of the PSU (the PFC circuit, etc...)
but the LM5033 still remains shut down through an optocoupler which senses something from the main board (the power button)
The GND for this 13V is on the primary side and is isolated from chassis
2) 16.75V on C85. it supplies some opamps and LM339 comparators on the secondary side. This supply is chassis referenced
3) 9.2V on the capacitor next to C85. This feeds an LM317 (SOIC package) that generates 5.18V
when the scope is powered on, the LM5033 comes on and generates the 6V and 12.6V.