I'm using the TD-1085/U timebase. I tried putting it in a verticle slot but I couldnt get a trace no matter what I did. Even pressing the beam finder.
The beam only unblanks if an amplifier or timebase is installed in the horizontal slot so to use the timebase in a vertical slot, install one of the vertical amplifiers into the horizontal slot.
You should still do this test. Based on what you describe below, the actual problem may be with the horizontal amplifier.
The delayed time base appears to be working, though I have no experience using it so I'm unsure. I checked the internal switches, there is a leaf switch behind the knob as well as a little switch on the board operated by the cam when switched from delayed / normal sweep. They appear to be in the correct positions / operating fine (I also beeped them out to be sure).
Usually the problem I see is that the delayed timebase cam switch breaks lose from the time/div control so when in delayed mode, the time/div control turns freely. I have a couple of 7B53As which failed this way. I repaired a 7B92A which had this problem and it generated a reverse sweep.
Now here is where things get kind of questionable and I am really unsure. On my timebase both sweep modes are "tied" together, that is both selectors are activated and both detents can be felt. When I pull the knob out, the sweep has a brightened portion and when turned clockwise only one set of detents are felt and the bright trace shrinks while the "normal" trace is still visible behind it. if I rotate it counter clockwise there is a single detent felt until I reach the \_/ shaped indicator position where the knob was when I pulled it out. Now it begins to decrease the normal timebase down as well as the delayed sweep. I can stop and turn clockwise again and the normal time base remains and the bright trace begins to "shrink" again. Also, when the knob is pulled out and I rotate the delay time multi I can "move" the bright (delayed?) trace left to right as I think one would expect. Is this all normal behavior?
That all sounds normal to me.
In order to see just the intensified zone, push in the time/div control knob. On the front of the control is says, "PULL FOR INTENS / PUSH FOR DELY'D". Now the time/div control sets the magnification and the delay time multiplier sets the position.
There is a tab on the clear plastic time/div ring which allows turning the main timebase counter clockwise while the time/div knob is pulled out.
I hope I explained this well enough.
You did fine. Dual delayed timebases are kind of weird until you gain some experience using them.
I do not have a Tektronix OS-245/P/U or any of the military plug-ins for it but I have the non-military non-ruggedized 7603 which it is based on as well as several other 7000 mainframes. I can probably answer any questions you have about this series of oscilloscopes.
These oscilloscopes are pretty old so suffer from various age related maladies but when working, they are very capable.