Apologies for the incomplete description of my tests, I will pay attention to that. Attached there are some images of the probe itself
and the compensation circuit.
I did a voltage divider test with the probe head and a 20M resistor, giving a 1:5 ratio in voltage drop, indicating the probe head is
indeed 100M.
I have also measured all the components in the compensation box with an LCR meter and the values overlap reasonably with the
specifications. I removed the compensation box-to-scope BNC connector because it was loose and could not be tightened
properly, which was affecting the reading if moved even minimally.
When I now use the probe, including the compensation box and the scope set at 1000x, to measure a 100 khz signal on the 50Ohm
dummy load, and compare it to the reading obtained using a standard Rigol probe at 10x, the result is still a bit off (image attached).
There seems to be a DC offset, as well as a phase shift, and the wave is not a clean sine.
When I try to adjust the compensation box, the effects are minimal. R5 does not change the amplitude at all, and tuning the other
components as described in the manual changes the waveshape in the right way but this effect is limited. The DC offset is not affected.
What other tests could I try?