Author Topic: Teardown : Keithley 2510 TEC SMU repair and calibration [complete]  (Read 21466 times)
Offline ManateeMafia
Re: Teardown : Keithley 2510 TEC SMU repair and calibration [complete]
« Reply #25 on: March 05, 2015, 02:52:05 am »
Your post reminded me that I bought a plcc adapter for my 2510 firmware.
I tried the secret menu but my rev A03 fw has no AT enable feature. I backed up my firmware and was going to flash the fw from your website until I noticed you were using the s file format.
Do you happen to have a bin file instead? I tried mot2bin.exe and it gave me a much larger file. I have not used this utility before today. Any help would be appreciated.
I will email you the A03 file when I get a chance.
« Reply #26 on: March 05, 2015, 05:24:14 am »
Never mind, I figured it out. I was able to use my programmer even though it wasn't immediately obvious. I did find another utility that would also work. It is 010 editor.
Offline TiNTopic starter
  • Country: ua
    • xDevs.com
« Reply #27 on: March 06, 2015, 03:52:39 pm »
I used WinHEX.
Binary was posted before in repository as well: http://dev.xdevs.com/projects/rnd/repository/show/Keithley/2510/firmware
 :bullshit:
Thanks for A03.
I am pretty positive that you may need recalibration, so beware.
« Last Edit: March 06, 2015, 03:54:25 pm by TiN »
« Reply #28 on: March 06, 2015, 11:58:43 pm »
You are correct, calibration constants appear to be lost. This is not a big deal, the only caveat is that I don't have a 1ohm 100W resistor. I have the rest in one form or another.
The serial number was also lost. I found your dump of the serial eeprom and it looks like the serial number is stored in the first set of bytes. I have ordered another adapter for my programmer.
I am looking at building a resistor array using some Riedon FPR2A-10RF1 or lower ohm value across a long heat sink. I have found a few enclosures and heat sinks I could mod. I have already done this for another test jig and it turned out fine.
  • Country: ua
    • xDevs.com
« Reply #29 on: March 07, 2015, 08:55:14 am »
You don't really need 100W rated resistor, SMU does not give that much power and it's pretty quick anyway.
I used 7W ones posted in thread above, worked well, and resistor was not even heat up.
« Reply #30 on: March 07, 2015, 12:33:40 pm »
That is good to know. I have a 1 ohm vhp-4 resistor but it has a max of 3A. I will find something a little beefier without the need for a another major project.
« Reply #31 on: March 10, 2015, 05:30:22 pm »
TiN,
Have you powered the 2510 through a GFCI outlet? So far, I have only had my 2510 trip the breaker. It happens almost immediately so it makes me think the SMPS inside the 2510 is causing the issue.
I only have one GFCI in the room and I will probably pick up a different adapter but I would like to know if you had any experience with this.
  • Country: ua
    • xDevs.com
« Reply #32 on: March 10, 2015, 05:55:32 pm »
I had to google what is that first. :)
No, i power all my gear from standard earthed outlets, which run into standard apartment breakers. Its 110VAC 60Hz here in Taiwan.
Offline gamalot
  • Correct my English
    • Youtube
« Reply #33 on: May 03, 2024, 12:23:51 pm »
I got my Keithley 2510 a few weeks ago and it has a slightly different main circuit board, the famous TF245 was replaced with a BI's resistor network and some chip resistors.
After upgrading the firmware from A03 to version A11, I was able to find the hidden "Autotune" menu item and set it to 2510-AT.
Thank you very much Ilya.
