David - thanks again, that all makes sense (although did you mean ’the roughly 0.7 volt Vbe of Q130’ rather than Q300?). I haven’t got a curve tracer, but the first thing I did was swap out Q300 with one from another SG503 awaiting investigation (which has a problem both with power supplies and another somewhere inside that TTL bird’s nest of counter logic - ugh). The change of transistor made no difference at all to the cutoff frequency.
I remade the joints around R140, both of which looked poor, but again sadly no difference to cutoff frequency. And everything tested OK with the correct DC resistances when I probed through.
Then I tried shunting R140 with a parallel 112 ohm resistor, thus bringing the total down to 45 ohms. Very slight difference to cutoff frequency. Here are my measured DC voltages in millivolts across R140 with and without the shunt (I’m purely concentrating on the low end of the 10 – 25 MHz range now). The voltages don’t rise any more beyond these maximum values as the frequency control is turned down further.
Frequency V (R140, 75 ohm) V (R140, shunted to 45 ohm)
12.0 232 377
11.5 286 448
11.0 410 640
10.9 429 682
10.8 469 779
10.7 529 943
10.6 660 1201
10.5 775 1289 (blinking, unlevelled)
10.4 1248 (blinking, unlevelled)
With that sudden ‘rush’ to the maximum voltage as the cutoff frequency is reached, I am coming back around to thinking that Q300 is performing correctly and doing its best, but it can’t do any more to rescue the collapsing oscillator circuit once it gets to around 1.9-1.248 = 0.65V Vce. In other words, Q300 bottoming-out is a symptom of the oscillator collapse, not the cause.
TERRA Operative - many thanks for the offer, although at the moment I can't think of what else to measure in a good reference unit that would give me an insight into my faulty unit.
jonpaul - thanks, I've cleaned the range wafer switches and the big S100 contacts many times, and I'm pretty convinced they are both OK - everything is completely consistent and stable as I twirl the range knob around. I've got the service manual. But your comment on the adjust / CAL procedure led me to fiddle again with the coil slug for this range (L112), partly in desperation. I had done this before, several months ago, and seen such crazy waveforms at the oscillator output that I got frightened and thought it best to leave well alone. But now I'm looking at the final output into 50 ohms with the famous 'precision' cable, (which I wasn't doing before), I can see a respectable looking sine wave even after I managed to move the slug to keep the oscillator alive at 9.90 MHz! I should have realised that all that heavy post-oscillator filtering is there for a reason.
But now of course I need to break out the spectrum analyser (the best I've got is FFT on a 200MHz Keysight DSOX, which I hope will be enough for this frequency range at least) and calibrate properly for harmonic distortion. I just hope it's not going to be too much of a trade-off between oscillator bandwidth and harmonic distortion, as per a previous comment.