Hello all,
thankyou very much for your attention on this problem, I really appreciate it.
I receipt your inputs and followed your suggestions.
Update:
Searching for a overload source for supply, I ended up desolder U701 (gpib controller TMS9914) that I felt heating up to 60°C.
Desoldering that IC had the effect to free the supply voltage to it regular value.
Testing resistance resistance between Vdd and Vss of the TMS i read 45ohm either polarity. Defenitely a bad value.
I think those are acceptable values.
- Following RolandK test i had these results:
Voltage across C753: 4.96V
Voltage across JM751: 0.8mV
Voltage input U553: 10.15V
Voltage Output U553: 4.85V - Checking factory issue, I can see a minimum gap between E100-101 and K101, there is a small air gap. Anyway I pushed E100 thoward S1.
Now U700 is correctly powered, U701 is missing because desoldered.
BUT.... the thing still doesnt work!
Oscillator of U700 is dead. I tried to takeout Y701, put two capacitors of 30pf and a cristal 12MHz but nothing happened.
Looking at signals, I saw Reset (pin 10) was high, so uC is held in reset state.
Look at the schematics, the chain that bring U700 to reset, became from U704.
Main board, after trying to establish communication, if doesn't have response send a pulse of about 80usec every 400msec.
In the image joined (sorry for the italian speach) you see pin 4 of U705B. Due to different tau introduced by R704-R708 and CR703, input of U705E is always high, and output always low.
So... why reset of U700 is high?
I tried to force reset pin down to gnd and the uC started to heat up to 60°C...
I don't know what happened to those ICs... but it seems that something bad happened there. Just a day turned on multimeter and this was the situation, apparently no event has trigger the damage.
So I'll try to find a couple of 87C11, I'm not sure I need to populate the GPIB controller too, but ok.. maybe I proceed one thing at time.
Thankyou very much coromonadalix for the firmware and Miti for the insight.
I think (or I hope) to have delimited the problem and I'll follow up on the results.
Cheers, Luca.