Hi All
There is a problem with front panel rotary knobs both volts/div as well as time/div
it is functioning randomly it doesnt go in sequence when rotated 10/5/2/1 after 1v/div instead of 500mv some times it switches back to 10V
Similarly the Time/div too behaves randomly
Any clues ?
Regards
Anand
Offline Jwalling
  • This is work?
Re: Tektronix TDS3012B
« Reply #1 on: August 03, 2019, 06:14:29 pm »
Dirt or some other contamination under the encoders, most likely. You have to remove them and clean the PCB pads and encoder contacts.
Jay
System error. Strike any user to continue.
  • Country: in
« Reply #2 on: August 07, 2019, 07:48:36 am »
hello jay
Thanks for the inputs
  • Country: in
« Reply #3 on: August 16, 2019, 04:57:12 am »
Hello jay
if you cut open the 4 places where the rotary knobs are sealed. after cleaning the contact what type of epoxy or glue do you recommend to reseal the knobs
Offline Jwalling
  • This is work?
« Reply #4 on: August 16, 2019, 09:13:45 am »
I use electronics grade RTV.
Jay
  • Country: in
« Reply #5 on: August 16, 2019, 02:16:42 pm »
hello jay
thanks
Anand
Offline Jwalling
  • This is work?
« Reply #6 on: August 16, 2019, 05:58:50 pm »
hello jay
thanks
Anand
Hi
Anand,
You're welcome. BTW, there's a "say thanks" button you can click on...  ;)
Jay
Offline warrjon
« Reply #7 on: August 16, 2019, 10:09:56 pm »
If cleaning the encoder does not work and you still have the issue RS Components has 12 step rotary encoders that you can replace the old ones with.
Offline Jwalling
  • This is work?
« Reply #8 on: August 17, 2019, 10:14:00 am »
I believe the encoders were custom made by Bourns for Tek. There's not much that can go wrong with them other than dirt or contamination. I've never seen the contacts on the encoder or the PCB wear out. If cleaning doesn't fix it, likely there's a electronics failure somewhere.