I got it from work; they wanted to get rid of it.
When turned on, it displayed the traces but no controls: all LEDs were lit and no reaction to any button/encoder, other than the power-up button.
Saw some oxidation on the flat cable connector solder pads on the Front panel board; touched up, no effect.
Eventually, after one of the start-ups, it worked. Then worked every time. Then, when left for an hour or so, started doing the same, this time not showing traces either. Then eventually, after many restarts, worked.
The voltages are all good, as per the table in the manual.
The troublehsooting chart points to main computer board, as the LEDs, at least some, or all, are lit.
The computer board has DS1230AB-120 Dallas chip, which, considering the scope is at least 30 years old, probably has its lithium source long dead. However it keeps the settings (mv/div, timing, the traces displayed), so maybe it is not dead yet.
Changing the computer module, as well as any other board, makes little sense due to the fact that the whole scope might sell for comparable money.
I am running continuous diagnostic until fail now. Not sure this will yield anything.
Anyone has experience with these?
PS. The test stopped, the message displayed was with all 5 tests "PASS". I switch it to run forever.