Here's a photo You requested.
LinkToday's results:
There are a bunch of opamps in the attenuation board. Most of them are dual opamps.
As i can find, some (or if not all) of them are controlled from device called 'daculator' (IMP I10412-01). U101 for ch1 & 2, U108 for ch3 & 4.
If I'm not mistaken, U117 are sullplying attenuator hybrids on channels 1 and 2, and U124 are supplying attenuator hybrids on channels 3 and 4.
U117 (LM358M) voltages (controlled by daculator U101):
pin1 is 5.51V, pin2 is -0.44V, pin3 is -0.44V, pin4 is -4.99V, pin5 is -0.44V, pin6 is -0.44V, pin7 is 5.51V, pin8 is 14.86V
These are supplying channels 1 and 2.
U124 (LM358M) voltages (controlled by daculator U108):
pin1 is 4.30V, pin2 is -0.90V, pin3 is -0.90V, pin4 is -4.94V, pin5 is -0.52V, pin6 is -0.52V, pin7 is 5.26V, pin8 is 14.86V
These are supplying channels 3 and 4.
These voltages are not changing when the scope glitches.
As we see, the voltages on channel 3 are worst.
Is it possible that due to the low supply of the channel 3, it glitches from time to time.
This discussion point me in that direction.
LinkThe problem is not temperature related (i think).
Tomorow I will pay more attention on the daculators.
This are the daculators according the Tek's documentation
A3U35 156–6224–00 IC,CONV:CMOS,D/A;12 BIT,V OUT,16 CHAN,SER IN,DACULATOR;I10412,SO28.300 80009 156–6224–00
LM358MMMAD1108PS: I'll be grateful if anyone has information or datasheet about the daculator: IMP I10412-01
Thanks.