Electronics > Repair

Tektronix TDS754D - acquisition board test fail

<< < (5/8) > >>

Rax:
Anyway, so I'm back at the drawing board and there's just a lot of scratching my head is all.

Rax:
A couple of notes on the rework:

* I think the .3mm solder I have is indeed crap. I switched to my default .8mm good quality solder, and I had to judiciously distribute each load across a few pins. Abundant rosin helped do that relatively easily. So, my solid belief quality tools and materials is everything is reinforced.
* The TT-65 and the SOJ tip made the desolder job very enjoyable. It's not a cheap solution, but I'm certain it cut down my time on that part of the job by a sizable percentage.

Rax:
As I evaluate the next steps here, I am trying to read the entirety of the error message. I am positive I read a while back a thread where this issue was discussed - can't recall if clear directions were presented - but can't find it now for the world.

I'd appreciate pointers on how to read the entire body of the individual errors.

TERRA Operative:
Ok, all good on the soldering, sounds like you got it all sorted. :)

To get the entire error messages, you'll need to read them out via GPIB. I also have no idea why Tektronix couldn't get the text to wrap properly on the scope display... :P


Here's how to read out the errors:

To read the first line of the error log via GPIB, send the command 'errlog:first?'
To read out subsequent lines, send the command 'errlog:next?' per line in sequence.
To clear the error log, send the command 'ERR CLEAR' via GPIB.


I would suggest clearing the error log first, then reboot the scope and manually run a full self test and SPC to generate a fresh set of error logs.

Rax:

--- Quote from: TERRA Operative on May 20, 2024, 12:09:36 am ---To read the first line of the error log via GPIB, send the command 'errlog:first?'
To read out subsequent lines, send the command 'errlog:next?' per line in sequence.
To clear the error log, send the command 'ERR CLEAR' via GPIB.


I would suggest clearing the error log first, then reboot the scope and manually run a full self test and SPC to generate a fresh set of error logs.

--- End quote ---

This is great, thank you. I was able to initially read an error from the dawn of time, but as trying out different things to read the last didn't work, I resorted to what you recommended and reset the error log. Then ran diagnostics again, and got this:

"SUN MAY 19 19:11:48 2024  ERROR: diagnostic test failure, digDataFormatDiag,  ERROR!!BYTE mode, in demux 200 ,i= 7 memBaseAdr 0x7380010=  data = 800, expectedData = 5500"

Looking forward to interpretations from those with experience on this instrument - as for me, I am headed now to do some reading.