Electronics > Repair
Tektronix TDS754D - acquisition board test fail
<< < (6/8) > >>
Rax:
BTW - not to jinx it, but I feel stronger and stronger that replacing those four SRAMs was called for, because there seems to be just one error now and those RAMs don't pop in there at all anymore.
TERRA Operative:
Yeah, looks like you had a few errors to deal with, at least you are making progress!

I'll have a search and think about your current error. My feeling is to either track down that address to a RAM chip and replace it as well (if not already done) or figure out what is controlling that RAM chip and stare intently at that.
I'd expect that if it was the chip passing the data to the RAM chips, it would be a periodic error across multiple chips rather than just one RAM chip in that case due to how the RAM is written to with bank switching etc.

One way to find the right RAM chip is to ground/apply 5V (have to double check which one) to the WE (Write Enable) pin one chip at a time and see what errors are thrown on self test due to that chip not being enabled to be written to.
If you get new errors on the same address range as your latest error, you found the chip.
Rax:
After running SPC, I got:

"SUN MAY 19 20:07:47 2024  ERROR: Internal adjustment range exceeded, ( 3 @   7464): DIGc gain = -0.400113, bal = 1.113079"

Bad news?...  :-\
TERRA Operative:
oooeerrrr.... Either bad cal (unlikely) or something is bad on that channel. Maybe as simple as a bad relay on the hybrid. It's not too hard to swap hybrids between channels, just need to be extra careful not to crack the ceramic PCB.
Relays aren't toooo difficult to replace, I have a video on Youtube about it.
youtube.com/watch?v=b0P1sJBomDQ

Is that the only error after clearing via GPIB?
Can you run SPC maybe 5 or 6 times to exercise the relays, then clear all errors, self test and SPC once more and list the entire error log here?
Rax:

--- Quote from: TERRA Operative on May 20, 2024, 05:43:45 am ---something is bad on that channel.

--- End quote ---
Would that be CH3? The error body is a bit cryptic to me.


--- Quote from: TERRA Operative on May 20, 2024, 05:43:45 am ---Is that the only error after clearing via GPIB?

--- End quote ---
Those two errors above, yes. The second error occurs if I run SPC.


--- Quote from: TERRA Operative on May 20, 2024, 05:43:45 am ---Can you run SPC maybe 5 or 6 times to exercise the relays, then clear all errors, self test and SPC once more and list the entire error log here?

--- End quote ---
Been doing a few of those, and the error seems pretty persistent.

I should also say, I don't feel there's anything very specifically wrong going on with CH3 - if that's the one found faulty by SPC. If anything, CH4 seems a bit choppier and there's a bit more freezing and vertical bars "noise." But they all work pretty well, their measurements are accurate as much as my other meters think. It just fails tests. Oh, and DPO mode doesn't work very well - there's a good amount of "video noise" - color pixelation ("snow") on the screen when that's enabled.
Navigation
Message Index
Next page
Previous page
There was an error while thanking
Thanking...

Go to full version
Powered by SMFPacks Advanced Attachments Uploader Mod