After switching pulse generators the first calibration step passes every time, strange too, as this generator appears to have a marginal rise time compared to the other. Anyway, the next step in the calibration procedure is with zero input, just a 50 Ohm termination, and, of course, the scope will not pass that test, ever. The scope measures the offset as follows:
Range Setting Ch 1 Ch 2
100V, 200V, 500V -340 V -220 V
10V, 20V, 50V -34 V -22 V
1 V, 2 V, 5 V -3.4 V - 2.2 V
500mV, 200 mV, 100mV -340 mV - 220 mV
I know the scaling of the offset like it does should give me a clue as to the defective circuit, I am guessing the DAC. Anyone care to make a guess. Thing is, the DAC appears to be part of a large integrated chip that would be nasty to replace even if I could get one. And since this is the only problem with this scope, the ac signal measure perfectly and look normal, it would have to be a failure unique to the DC level adjust. I was hoping for a reference voltage failure somewhere. Any ideal will be greatly appreciated!