Hello, there are two topics related to TM503 power module and specific voltage calibrator PG506 section.
In the past, I've been calibrating legacy 2445 and 2465 tektronix oscilloscope where the key sections are PG506, SG503-SG504 and TG501 installed in TM50x power supply set. I was quite happy with this modular approach so acquired through times total of four TM503 (no fan) and many other sections (DC503A, PS50A, DM501A, FG504, AA501...) used in lab testing plus another PG506 from Israel, the first one from Greece. Lately I've acquired TDS540C along with some repairs so was looking on specific testing gear, it seems PG506 is quite useful.
Last week I've decided to run again both PG506 and suddenly they failed each with different symptoms. I'm pretty certain a possible root cause could be one TM503 because these PG506 used to run fine since 2014. The 1st PG506 is severely damaged but I was able to repair partially the 2nd PG506.
After spending lot of time reviewing the service manual of PG506 and TM503, it seems the PG506 uses two transistors found inside the TM503. The schematics of the PG506 shows a complex relay switching with different modes involving many local power supplies plus an earth ground and a floating ground.
My first question about TM503: do you recommend for this legacy units to recondition capacitor or inspect to make sure no strange transient could transmitted to the tektronix section ?
My second question about PG506:after looking the schematics, I've discovered in fact different models through time with PCB variants where I guess tektronix improved or solved known failures. The 1st generation with S/N below B040000, the 2nd generation from S/N above B040000 then the PG506A. Unfortunately I did purchase on eBay B03xxx so probably the one with most failures.
What is your return of experience on 1st generation PG506, are they reliable and worth repair or better look for PG506A ?
My 3rd question about PG506: I found finally one tantalum 16.5Vdc filtering capacitor C170 to be shorted so I've replaced both C170 and C180 from the other unit. However I notice both units have very old capacitors so wonder if it is recommended to recap plus what could be the root cause that the Standard Amplitude output to have lot of noise (see attached picture).
I've probed down to Q255 output, the signal is clean so there must be an internal power supply filter somewhere but there are so many power supplies (+/-16.5V, 5V, 18V, 120V, -72V, 14V...).
Another way to ask or frame my questions:
- is is there any known recurrent specific failures of the PG506's versus the TM503 because most of the sections in my four TM503 always worked fine
- what check or recondition is suggested on old TM503's
The key point being if these legendary PG506 are still worth to be used in labs and worth being repaired.
Thank you, Albert