Hi all,
Thought I'd document a repair I did recently to a bit of an "unknown" brand in the world of metrology. It seems there's little information out there on these (dare I say) quite fantastic multimeters. With their ability to go all the way up to 10kV and 30A with very high resolution they're imo very useful meters. I believe the datasheet was something like 18ppm/yr.
First of all, I'm sure with a quick Google you'll notice this is a 7.5 digit meter, which is true (kinda...). Later I'll describe how you can "unlock" this to an 8.5 digit meter which doesn't seem to be listed in any manuals anywhere. This is a true last digit and not just some approximation using some foolery from what I can tell.
I bought this particular unit off eBay for a couple £100 broken without much description of the exact fault, but for the price I thought I'd give it a go. It of course lacked a recent cal (2012) and the Cal key was also missing (a 2.5mm shorted jack) but this could overlooked. I gave it a quick test and all appeared to be working well comparing it to my MF calibrator so wasn't sure what the problem was (and was worrying I'd be unable to justify tearing it apart!)
Until I checked the capacitance range. Completely dead.
I've attached some images, the back "blocks" are of similar construction to other Time Electronics gear - I'm tempted to x-ray the A/D module as it's potted, but the AC module was not!
While I was in there, I dumped the content of any eeproms and ICs I could find. There is a battery, but from the manual, this appears to be RTC only and not backing up any vital cal data and the like. The rest of the digital board is so so, bunch of logic, power supply bits and GPIB circuitry. There was a curious looking burnt resistor, but this was for the frequency and unrelated. I changed it for a 0.5W because it gave me the chills looking all brown even if it was still in tolerance.
On the analog side we have two blocks on separate PCBs, an amplifier on its own board, some optos, attenuator sections and capacitance/frequency/diode sections.
As I'm sure many of you will notice, there are VPG resistors everywhere. Underneath the board is a large shunt resistor (not pictured) for the high current shunt.
Skipping over the boring diagnostic processes, the capacitance section is at the top left of the board. It appears that REG3 was dead and so was TR30. I suspect one took the other out, but as to who was the culprit, we will never know...
Changing these got the capacitance range up and running again, so now it'll be left to settle for a while, then will receive a full cal!
On to the fun unlock, it appears that with the Cal key in, I was able to hold MENU while the meter booted up. This enabled a bunch of options in the menu. The one to look for here is "MAX RESOLUTION". This enabled "8 DIGIT RESOLUTION" under "N DIGITS" button. Make sure you change this and save your changes before exiting!
I believe this is a "real" and not averaged digit, as you can change the LSD (8th digit) on all the ranges completely, and the sample time does not change. This is no solartron sample time either, this meter does many samples a second!
If anyone is interested in manuals, or has anything else to add, feel free to PM me! I may add to this post if I think of anything else and will probably x-ray the potted section.