I am Trying to fix a Dick smith Q-1802 oscilloscope has also been rebranded to a Toptronic T200 and digitor Q-1802 pintek PS-200 oscilloscope.
It Started off with a dead primary on the transformer which i think was undersized for the scope in the first place .


PS 20-60mhz (1).pdf (1211.21 kB - downloaded 22 times.)" alt="" class="bbc_img" />
DSCN6983.JPG (107.41 kB. 800x600 - viewed 99 times.)
I replaced the dead transformer with a similar scrapped scope transformer and second for the low voltage supply as i didn't have a correct replacement
Pretty much getting correct voltages and the scope lights up but the problem is i have no vertical position control on both channels and inputting a waveform result not correct .
No parts look damaged i replaced the 4011's and 7474's so far but it might be a transistor .
I suspect the High Negative is discharging next to any close wiring caused the problem it should of had another jacket of plastic over it as in later models , i can fix that but finding the problem on the vertical side needs looking into ..
Getting pretty much no trace movement from the vertical position controls ,no problems on the horizontal side .